5.4.3.2. SETUP  RNGE  MODE  IND: CONFIGURING THE T200 ANALYZER FOR
INDEPENDENT RANGE MODE
86
Teledyne API – T200 NO/NO
The independent range mode (IND) assigns the three NO
to individual analog output channels. In IND range mode the RANGE test function
displayed on the front panel will then be replaced by three separate functions:
Table 5-1:
IND Mode Analog Output Assignments
TEST
CONCENTRATION
FUNCTION
RANGE1
RANGE2
RANGE3
Each can be configured with a different reporting range upper limit and analog signal
span:
EXAMPLE:
NO
Concentration – RANGE1 Set for 0-800 ppb & output A1 set for 0-10 VDC
x
NO Concentration – RANGE2 Set for 0-200 ppb & output A2 set for 0-5 VDC
NO
Concentration – RANGE3 Set for 0-400 ppb & output A3 set for 0-5 VDC
2
Setting analog range limits to different values does not affect the instrument‟s
calibration.