Teledyne API – T200 NO/NO
WARNING
RCELL PRESS WARN
RCELL TEMP WARNING
REAR BOARD NOT DET
RELAY BOARD WARN
SAMPLE FLOW WARN
SYSTEM RESET
Note
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
FAULT CONDITION
Sample Pressure is <15 in-Hg or >
35 in-Hg
Normally 29.92 in-Hg at sea level
decreasing at 1 in-Hg per 1000 ft of
altitude (with no flow – pump
disconnected).
RCELL temperature is < 45C or >
55C.
Motherboard not detected on
power up.
The CPU cannot communicate with
the Relay Board.
Sample flow rate is < 350 cc/min or >
600 cc/min.
The computer has rebooted.
A failure of the analyzer's CPU, motherboard or power supplies can result in any
or ALL of the above messages.
POSSIBLE CAUSES
If Sample Pressure is < 15 in-HG:
Blocked Particulate Filter
Blocked Sample Inlet/Gas Line
Failed Pressure Senor/circuitry
If Sample Pressure is > 35 in-HG:
Bad Pressure Sensor/circuitry
Pressure too high at Sample Inlet.
Heater configured for wrong voltage type
Failed RCELL Temperature Sensor
Relay controlling the heater is not working
Failed Relay Board
2
I
C Bus
This WARNING only appears on Serial I/O COM Port(s) Front
Panel Display will be frozen, blank or will not respond.
Failure of Motherboard
2
I
C Bus failure
Failed Relay Board
Loose connectors/wiring
Failed Sample Pump
Blocked Sample Inlet/Gas Line
Dirty Particulate Filter
Leak downstream of RCELL Critical Flow Orifice
Failed Sample Pressure Sensor
Failed Vacuum Pressure Sensor
This message occurs at power on.
If it is confirmed that power has not been interrupted:
Failed +5 VDC power
Fatal Error caused software to restart
Loose connector/wiring
Troubleshooting & Service