Figure 5-8: Alternative Setup Using 250Ω Resistor For Checking Current Output Signal Levels; Table 5-8: Current Loop Output Check - Teledyne T200 Operation Manual

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Teledyne API – T200 NO/NO
Alternative Setup Using 250Ω Resistor for Checking Current Output Signal Levels
Figure 5-8:
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
An alternate method for measuring the output of the Current Loop converter is to
connect a 250 ohm 1% resistor across the current loop output in lieu of the current
meter (see Figure 3-8 for pin assignments and diagram of the analog output connector).
This allows the use of a voltmeter connected across the resistor to measure converter
output as VDC or mVDC.
V OUT +
V OUT -
ANALYZER
In this case, follow the procedure above but adjust the output for the following values:
Table 5-8:
Current Loop Output Check
Voltage across
% FS
Resistor for 2-20 mA
0
100
V
Meter
+DC
Gnd
V IN +
250 Ω
V IN -
Resistor for 4-20 mA
500 mVDC
5000 mVDC
Setup Mode Menus
Volt
Recording
Device
Voltage across
1000 mVDC
5000 mVDC
115