Figure 3-24: Gas Line Connection W/Ambient Zero/Pressurized Span Valves (Opt 50B) - Teledyne T200 Operation Manual

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Getting Started
Figure 3-24:
58
Teledyne API – T200 NO/NO
VENT
here if input
is pressurized
Zero Air Scrubber
Gas Line Connection w/Ambient Zero/Pressurized Span Valves (OPT 50B)
S
G
S
AMPLE
AS
OURCE
Attach a sample inlet line to the SAMPLE inlet fitting.
Use PTFE tubing; minimum O.D ¼".
Sample Gas pressure must equal ambient atmospheric pressure (29.92 in-Hg).
In applications where the sample gas is received from a pressurized manifold, a
vent must be placed on the sample gas line. This vent line must be:
no more than 10 meters long
vented outside the shelter or immediate area surrounding the instrument
C
G
S
ALIBRATION
AS
OURCES
Attach a gas line from the pressurized source of calibration gas (e.g. a
SPAN GAS
bottle of NISTSRM gas) to the SPAN1 inlet. Use PTFE tubing, minimum
O.D ¼".
(the dual-stage zero Air Scrubber makes zero air)
ZERO AIR
V
ENTING
Attach a line to the SPAN2/VENT outlet. It should be:
¼" PTFE tubing
vented outside the enclosure
not less than 2 meters in length
not greater than 10 meters in length
SAMPLE Gas
Source
SAMPLE
PUMP
EXHAUST
SPAN1
SPAN2/
ZERO AIR
/NO
Analyzer Operation Manual
2
X
Chassis
VENT
06858E DCN7057