OPTICAL TEST (OTEST) CONTROL: When the CPU sets this control signal to
high (ON) the optical test feature is initiated (see Section 8.3).
The OTEST can be initiated by following the procedure in 12.7.12.1, or by setting
the SIGNAL I/O Function OPTIC_TEST to ON.
PMT PREAMPLIFIER RANGE CONTROL: The CPU uses this control switch the
instrument between its LOW and HIGH physical ranges (see Section 5.4.1).
The instrument can be forced into its HIGH physical range setting the SIGNAL I/O
function PREAMP_RANGE_HI to ON.
O
GEN STATUS: The CPU uses this control signal to turn the O
3
by setting it to HIGH/LOW respectively. The CPU turns OFF the O
is if there is no or low air flow to it as measured by the O
instrument has been turned off for more than 30 minutes.
The O
generator can be manually turned ON/OFF by using the OZONE
3
GENERATOR OVERIDE feature (See Section 12.7.15.1) or by setting the SIGNAL
I/O function O3GEN_STATUS to ON or OFF.
Any I/O signals changed while in the signal I/O menu will remain in effect ONLY
until signal I/O menu is exited.
The analyzer regains control of these signals upon exit and returns them to their
normal value/setting.
2
I
C is a two-way, clocked, bi-directional digital serial I/O bus that is used widely in
commercial and consumer electronic systems.
converts data and control signals from the PC-104 bus to I
fed to the relay board, optional analog input board and valve driver board circuitry.
This circuit monitors the +5V power supply during start-up and sets the analog outputs,
external digital I/O ports, and I
instrument software can establish control.
A transceiver on the Motherboard
2
C circuitry to specific values until the CPU boots and the
Principles of Operation
generator ON/OFF
3
generator if there
3
flow sensor or if the
3
2
C format. The data is then