Setup Mode Menus
5.9.3.2. CALIBRATION OF THE ANALOG OUTPUTS
5.9.3.3. ENABLING OR DISABLING THE AUTOCAL FOR AN INDIVIDUAL ANALOG OUTPUT
NOTE:
ANALOG OUTPUTS
configured for 0.1V full
scale should always be
calibrated manually.
Toggle this button
to turn AUTO CAL
ON or OFF
(OFF = manual
calibration mode).
108
Teledyne API – T200 NO/NO
In its default mode, the instrument is configured for automatic calibration of all
channels, which is useful for clearing any analog calibration warnings associated with
channels that will not be used or connected to any input or recording device, e.g., data
logger.
Manual calibration should be used for the 0.1V range or in cases where the outputs must
be closely matched to the characteristics of the recording device. The AUTOCAL
feature must be disabled first for manual calibration.
To enable or disable the AutoCal feature for an individual analog output, elect the
ANALOG I/O CONFIGURATION submenu (see Figure 5-5) then press:
AIO CONFIGURATION SUBMENU
DIAG
PREV NEXT
DIAG AIO
SET>
Continue pressing SET> until you reach the
DIAG AIO
<SET SET>
DIAG AIO
SET>
Continue pressing SET> until ...
DIAG AIO
<SET SET>
DIAG AIO
ON
DIAG AIO
OFF
From the
ANALOG I/O CONFIGURATION
ENTR
AOUTS CALIBRATED: NO
CAL
output to be configured
CONC_OUT_2: 5V, OVR, NOCAL
EDIT
CONC_OUT_2: RANGE: 5V
EDIT
CONC_OUT_2: AUTO CAL.:ON
EDIT
CONC_OUT_2: AUTO CAL.:ON
ENTR EXIT
CONC_OUT_2: AUTO CAL.:OFF
ENTR EXIT
/NO
Analyzer Operation Manual
2
X
EXIT
EXIT
EXIT
EXIT
EXIT
ENTR accepts
the new setting.
EXIT ignores the
new setting.
06858E DCN7057