Inet (Ethernet); Com1[Com2] (Mode, Baude Rate And Test Port); Setup  Vars: Variables Setup And Definition; Figure 5-3. Comm- Machine Id - Teledyne T200 Operation Manual

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Teledyne API – T200 NO/NO
Toggle to cycle
through the available
character set: 0-9

5.7.2. INET (ETHERNET)

5.7.3. COM1[COM2] (MODE, BAUDE RATE AND TEST PORT)

5.8. SETUP  VARS: VARIABLES SETUP AND DEFINITION
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
SETUP X.X
COMMUNICATIONS MENU
INET
COM1
ID
SETUP X.
MACHINE ID: 200 ID
0
2
0
Figure 5-3.
The ID can be any unique 4-digit number and can also be used to identify analyzers in
other ways (e.g. location numbers, company asset number, etc.)
Use SETUP>COMM>INET to configure Ethernet communications, whether manually
or via DHCP. Please see Section 6.5.2 for configuration details.
Use the SETUP>COMM>COM1[COM2] menus to:
configure communication modes (Section 6.2.1)
view/set the baud rate (Section 6.2.2)
test the connections of the com ports (Section 6.2.3)
Configuring COM1 or COM2 requires setting the DCE DTE switch on the rear panel.
Section 6.1 provides DCE DTE information.
Through the SETUP>MORE>VARS menu there are several user-adjustable software
variables that define certain operational parameters. Usually, these variables are
automatically set by the instrument‟s firmware, but can be manually re-defined using the
VARS menu.
Table 5-3 lists all variables that are available within the 818 password protected level.
See Appendix A2 for the T200 variables that are accessible through the remote interface.
COM2
0
ENTR EXIT
COMM– Machine ID
Setup Mode Menus
EXIT
ENTR accepts the new
settings
EXIT ignores the new
settings
99