3.4. STARTUP, FUNCTIONAL CHECKS, AND INITIAL
CALIBRATION

3.4.1. START UP

3.4.2. WARNING MESSAGES

64
Teledyne API – T200 NO/NO
If the presence of ozone is detected at any time, power down the instrument
and contact Teledyne API Technical Support as soon as possible:
800-324-5190 or email:
We recommend that you read Section 13 to become familiar with the T200 principles of
operation. For information on navigating the analyzer‟s software menus, see Appendix
A.
After making the electrical and pneumatic connections, run an initial functional check.
Turn on the instrument. The pump and exhaust fan should start immediately. The
display will show a splash screen and other information during the initialization process
while the CPU loads the operating system, the firmware, and the configuration data.
The analyzer should automatically switch to Sample Mode after completing the boot-up
sequence and start monitoring the gas. However, there is an approximately one hour
warm-up period before reliable gas measurements can be taken. During the warm-up
period, the front panel display may show messages in the parameters (Param) field.
Because internal temperatures and other conditions may be outside the specified limits
during the analyzer‟s warm-up period, the software will suppress most warning
conditions for 30 minutes after power up. If warning messages persist after the 30
minutes warm up period is over, investigate their cause using the troubleshooting
guidelines in Section 12.1.
To view and clear warning messages, press:
CAUTION!
sda_techsupport@teledyne.com
/NO
Analyzer Operation Manual
2
X
06858E DCN7057