Teledyne T200 Operation Manual page 91

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Teledyne API – T200 NO/NO
5.4.3.3. SETUP  RNGE  MODE  AUTO: CONFIGURING THE T200 ANALYZER FOR AUTO
RANGE MODE
I
MPORTANT
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
In AUTO range mode, the analyzer automatically switches the reporting range between
two user-defined ranges (LOW and HIGH). The same low and high span settings are
applied equally to NO, NO
The unit will switch from LOW range to HIGH range when either the NO, or NO
concentration exceeds 98% of the low range span.
The unit will return from HIGH range back to LOW range once both the NO and
NO
concentrations fall below 75% of the low range span.
X
The LOW & HIGH ranges referred to here are NOT the same as the low & high
physical ranges referred to in Section 5.4.1.
Also the RANGE test function displayed on the front panel will be replaced by two
separate functions:
RANGE1: The LOW range setting for all analog outputs
RANGE2: The HIGH range setting for all analog outputs
The LOW/HIGH range status is also reported through the external, digital status bits
(Section 3.3.1.4).
and NO
readings.