Table 5-6: Analog Output Voltage Range Min/Max - Teledyne T200 Operation Manual

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Teledyne API – T200 NO/NO
5.9.3.1. ANALOG OUTPUT VOLTAGE / CURRENT RANGE SELECTION

Table 5-6: Analog Output Voltage Range Min/Max

RANGE NAME
0.1V
1V
5V
10V
The default offset for all VDC ranges is 0-5 VDC.
CURR
While these are the physical limits of the current loop modules, typical applications use 2-20 mA or 4-20 mA for the lower and upper
limits. Please specify desired range when ordering this option.
The default offset for all current ranges is 0 mA.
These buttons
set the signal
level and type
of the selected
channel.
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
In its standard configuration the analog outputs is set to output a 0 – 5 VDC signals.
Several other output ranges are available (see Table 5-6). Each range is usable from -
5% to +5% of the rated span.
RANGE SPAN
0-100 mVDC
0-1 VDC
0-5 VDC
0-10 VDC
0-20 mA
To change the output type and range, select the ANALOG I/O CONFIGURATION
submenu (see Figure 5-5) then press:
From the
AIO CONFIGURATION SUBMENU
DIAG
ANALOG I/O CONFIGURATION
PREV NEXT
DIAG AIO
AOUTS CALIBRATED: NO
CAL
SET>
Continue pressing SET> until you reach the
output to be configured
DIAG AIO
CONC_OUT_2: 5V, OVR, NOCAL
<SET SET>
EDIT
DIAG AIO
CONC_OUT_2 RANGE: 5V
<SET SET>
EDIT
DIAG AIO
CONC_OUT_2: RANGE: 5V
0.1V
1V
5V
10V CURR
MINIMUM OUTPUT
-5 mVDC
-0.05 VDC
-0.25 VDC
-0.5 VDC
0 mA
EXIT
ENTR
EXIT
EXIT
EXIT
ENTR EXIT
Setup Mode Menus
MAXIMUM OUTPUT
105 mVDC
1.05 VDC
5.25 VDC
10.5 VDC
20 mA
Pressing ENTR records
the new setting and
returns to the previous
menu.
Pressing EXIT ignores the
new setting and returns to
the previous menu.
107

Advertisement

Table of Contents
loading

This manual is also suitable for:

T200uT201T265T200u-noyT200up

Table of Contents