Setup Mode Menus
5.9.3.5. AUTOMATIC INDIVIDUAL CALIBRATION OF THE ANALOG OUTPUTS
110
Teledyne API – T200 NO/NO
Use the AUTO CAL feature to initiate an automatic calibration for an individual analog
output;
access
the
(SETUP>MORE>DIAG>PASSWORD>NEXT . . . or see Figure 5-5) then press:
AIO CONFIGURATION SUBMENU
DIAG
ANALOG I/O CONFIGURATION
PREV NEXT
DIAG AIO
AOUTS CALIBRATED: NO
CAL
SET>
Continue pressing SET> until you reach the
output to be configured
DIAG AIO
CONC_OUT_2: 5V, CONC2, NOCAL
<SET SET>
EDIT
DIAG AIO
CONC_OUT_2: RANGE: 5V
EDIT
SET>
Continue pressing SET> until ...
DIAG AIO
CONC_OUT_2: CALIBRATED:NO
<SET SET>
CAL
DIAG AIO
AUTO CALIBRATING CONC_OUT_2
DIAG AIO
CONC_OUT_2: CALIBRATED: YES
<SET SET> CAL
ANALOG
I/O
From the
ENTR
/NO
Analyzer Operation Manual
2
X
CONFIGURATION
EXIT
EXIT
EXIT
EXIT
EXIT
EXIT
06858E DCN7057
submenu