Teledyne T200 Operation Manual page 74

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Getting Started
Analyzer continues to
cycle through NO
NO, and NO
measurements
throughout this
procedure.
The SPAN button now
appears during the transition
from zero to span.
You may see both buttons.
If either the ZERO or SPAN
buttons fail to appear see the
Troubleshooting Section of
this manual.
Note
72
Teledyne API – T200 NO/NO
SAMPLE
< TST
TST >
,
x
2
SAMPLE
< TST TST >
Allow zero gas to enter the sample port
SAMPLE
< TST TST >
M-P CAL
<TST TST>
M-P CAL
<TST TST>
Allow span gas to enter the sample port
SAMPLE
< TST TST >
M-P CAL
NOX STB= XXX.X PPB
<TST TST> ZERO
M-P CAL
<TST TST>
M-P CAL
<TST TST> ENTR
The T200 Analyzer is now ready for operation.
Once you have completed the above set-up procedures, please fill out the
Quality Questionnaire that was shipped with your unit and return it to Teledyne
API. This information is vital to our efforts in continuously improving our
service and our products. THANK YOU.
RANGE=500.0 PPB
NOX= XXXX
CAL
SETUP
Toggle TST> button until ...
NOX=XXX.X
NOX STB= XXX.X PPB
CAL
SETUP
at the rear of the analyzer.
NOX=XXX.X
NOX STB= XXX.X PPB
SETUP
CAL
NOX=XXX.X
NOX STB= XXX.X PPB
ZERO
CONC
EXIT
NOX=X.XXX
NOX STB= XXX.X PPB
CONC
ENTR
at the rear of the analyzer.
NOX=XXX.X
NOX STB= XXX.X PPB
SETUP
CAL
NOX=X.XXX
CONC
EXIT
SPAN
NOX=X.XXX
NOX STB= XXX.X PPB
CONC
EXIT
ENTR
NOX=X.XXX
NOX STB= XXX.X PPB
CONC
EXIT
/NO
Analyzer Operation Manual
2
X
Set the Display to show
the NOX STB test
function.
This function calculates
the stability of the NO/NO
measurement
Wait until NOX STB
falls below 0.5 ppB.
This may take several
minutes.
Press ENTR to changes
the OFFSET & SLOPE
values for both the NO
and NO
measurements.
x
Press EXIT to leave the
calibration unchanged and
return to the previous
menu.
EXIT
Wait until NOX STB
falls below 0.5 PPB.
This may take several
minutes.
Press ENTR to changes
the OFFSET & SLOPE
values for both the NO
and NO
measurements.
x
Press EXIT to leave the
calibration unchanged and
return to the previous
menu.
EXIT at this point
returns to the
SAMPLE menu.
06858E DCN7057