Teledyne T200 Operation Manual page 120

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Setup Mode Menus
5.9.3.10. AIN CALIBRATION
DIAG AIO
Perform
Troubleshooting or call
Teledyne API's
Customer Service
118
Teledyne API – T200 NO/NO
This is the submenu to conduct a calibration of the T200 analyzer‟s analog inputs. This
calibration should only be necessary after major repair such as a replacement of CPU,
motherboard or power supplies.
To perform an analog input calibration, select the ANALOG I/O CONFIGURATION
submenu (see Figure 5-5) then press:
AIO CONFIGURATION SUBMENU
DIAG
PREV NEXT
DIAG AIO
<SET
SET>
Continue pressing SET> until you reach the
DIAG AIO
<SET
DIAG AIO
Firmware automatically performs a zero point calibration
of the Motherboard's analog Inputs
DIAG AIO
Firmware automatically performs a span point calibration
of the Motherboard's analog Inputs
A/D CALIBRATION ERROR
From the
ANALOG I/O CONFIGURATION
ENTR
AOUTS CALIBRATED: NO
CAL
output to be configured
AIN CALIBRATED:NO
CAL
CALIBRATING A/D ZERO
CALIBRATING A/D SPAN
DIAG AIO
EXIT
DIAG AIO
<SET
/NO
Analyzer Operation Manual
2
X
EXIT
EXIT
EXIT
AIN CALIBRATED: YES
AIN CALIBRATED:NO
CAL
06858E DCN7057
EXIT
EXIT