Figure 3-8: Analog Output Connector; Table 3-5: Analog Output Pin Assignments - Teledyne T200 Operation Manual

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Teledyne API – T200 NO/NO

Table 3-5: Analog Output Pin Assignments

PIN
ANALOG OUTPUT
1
2
3
4
3
4
7
8
3.3.1.4. CURRENT LOOP ANALOG OUTPUTS (OPTION 41) SETUP
Servicing or handling of circuit components requires electrostatic discharge
protection, i.e. ESD grounding straps, mats and containers. Failure to use ESD
protection when working with electronic assemblies will void the instrument
warranty. For information on preventing ESD damage, refer to the manual,
Fundamentals of ESD, PN 04786, which can be downloaded from our website at
http://www.teledyne-api.com
Manuals section.
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
ANALOG OUT
A1
+
-
+
Figure 3-8:
A1
NO
A2
NO Concentration
A3
NO
1
A4
TEST CHANNEL
To change the settings for the analog output channels, see Section 5.9.2.
If your analyzer had this option installed at the factory, there are no further connectons
to be made. Otherwise, it can be installed as a retrofit for each of the analog outputs.
This option converts the DC voltage analog output to a current signal with 0-20 mA
output current, which can be scaled to any set of limits within that 0-20 mA range.
However, most current loop applications call for either 2-20 mA or 4-20 mA range. All
current loop outputs have a +5% over-range. Ranges with the lower limit set to more
than 1 mA (e.g., 2-20 or 4-20 mA) also have a -5% under-range.
Figure 3-9 provides installation instructions and illustrates a sample configuration of one
current output combined with two voltage outputs . Next are instructions for converting
current loop analog outputs to standard 0-to-5 VDC outputs. To calibrate or adjust these
outputs see Section 5.9.3.7.
CAUTION – AVOID INVALIDATING WARRANTY
A2
A3
A4
-
+
-
+
Analog Output Connector
STANDARD
SIGNAL
VOLTAGE OUTPUT
Concentration
x
Concentration
2
under Help Center > Product Manuals in the Special
Getting Started
-
CURRENT
LOOP OPTION
V Out
I Out +
Ground
V Out
I Out +
Ground
V Out
I Out +
Ground
V Out
Not Available
Ground
Not Available
I Out -
I Out -
I Out -
35