Test Chan Output (Selecting A Test Channel Function For Output A4) - Teledyne T200 Operation Manual

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Setup Mode Menus
5.9.4. TEST CHAN OUTPUT (SELECTING A TEST CHANNEL FUNCTION
FOR OUTPUT A4)
Table 5-9: Test Channels Functions available on the T200's Analog Output
TEST CHANNEL
NONE
PMT DETECTOR
OZONE FLOW
SAMPLE FLOW
SAMPLE PRESSURE
RCELL PRESSURE
RCELL TEMP
MANIFOLD TEMP
IZS TEMP
CONV TEMP
PMT TEMP
BOX TEMP
HVPS VOLTAGE
120
Teledyne API – T200 NO/NO
The test functions available to be reported are listed in Table 5-9:
DESCRIPTION
The output of the PMT detector
converted to a 0 to 5 VDC scale.
The flow rate of O
3
as measured by the O
The calculated flow rate for sample gas
through the analyzer.
The pressure of the sample gas
measured upstream of the Auto Zero
Valve.
The pressure of gas inside the reaction
cell of the sensor module.
The temperature of gas inside the
reaction cell of the sensor module.
Not used in the Model T200.
The temperature of the permeation tube
oven of the optional internal span gas
generator.
The temperature NO
The temperature inside PMT.
The temperature inside the T200's
chassis.
Represents the output voltage of the
PMT's high voltage power supply.
Once a function is selected, the instrument not only begins to output a signal on the
analog output, but also adds TEST to the list of test functions viewable via the front
panel display.
TEST CHANNEL IS TURNED OFF
through the analyzer
flow sensor.
3
 NO converter.
2
/NO
Analyzer Operation Manual
2
X
ZERO
FULL SCALE
0 mV
5000 mV
3
0 cm
/min
1000 cm
3
0 cm
/min
1000 cm
0 Hg-In-A
40 "Hg-In-A
0 Hg-In-A
40 Hg-In-A
0 C
70 C
0 C
70 C
0 C
500 C
0 C
50 C
0 C
70 C
0 mV
5000 mV
06858E DCN7057
1
3
/min
3
/min
1