Figure 12-5: Typical Set Up Of Status Output Test - Teledyne T200 Operation Manual

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Teledyne API – T200 NO/NO
12.7.6.3. STATUS OUTPUTS
SYSTEM_OK
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
1
2
3
Figure 12-5:
Typical Set Up of Status Output Test
To test the status output electronics:
1. Connect a jumper between the "D" pin and the "" pin on the status output
connector.
2. Connect a 1000 ohm resistor between the "+" pin and the pin for the status output
that is being tested.
3. Connect a voltmeter between the "" pin and the pin of the output being tested.
4. Under the DIAG Signal I/O menu (see Section 12.1.3), scroll through the inputs
and outputs until you get to the output in question.
5. Alternately, turn on and off the output noting the voltage on the voltmeter.
It should vary between 0 volts for ON and 5 volts for OFF.
Table 12-10:
Status Outputs Check
PIN (LEFT TO RIGHT)
1
2
3
4
5
6
7
8
V
+DC
Gnd
4
5
6
7
8
D
1000 Ω
STATUS
ST_SYSTEM_OK
ST_CONC_VALID
ST_HIGH_RANGE
ST_ZERO_CAL
ST_SPAN_CAL
ST_DIAG_MODE
Not Used on T200
ST_O2_CAL