Teledyne API – T200 NO/NO
V
ENTING
To prevent back diffusion and pressure effects, both the span gas and zero air supply
lines should be:
vented outside the enclosure
minimum OD ¼"
not less than 2 meters in length
not greater than 10 meters in length
E
O
XHAUST
UTLET
Attach an exhaust line to the EXHAUST outlet fitting. The exhaust line should be:
of PTFE tubing; minimum OD ¼"
maximum of 10 meters long
vented outside the T200 analyzer's enclosure
Once the appropriate pneumatic connections have been made, check all
pneumatic fittings for leaks using the procedures defined in Sections 11.3.12
(or 11.3.12.2 for detailed check if leak suspected).
/NO
Analyzer Operation Manual
2