Figure 3-26: Rear Panel Layout With Internal Span Source (Izs) Opt 50G - Teledyne T200 Operation Manual

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Teledyne API – T200 NO/NO
Figure 3-26:
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
Rear Panel Layout with Internal Span Source (IZS) OPT 50G
I
S
G
NTERNAL
PAN
AS
The primary component of the internal span option is a permeation tube containing
liquid NO
As zero air is passed over a permeable membrane on the end of the tube,
2.
molecules of NO
slowly pass through the membrane mixing with the zero air.
2
The resulting concentration of the NO
size of the membrane (the larger the area of the membrane, the more permeation
occurs)
temperature of the NO
increases the pressure inside the tube, thereby increasing the rate of permeation)
flow rate of the zero air (if the previous two variables are constant, the permeation
rate of the NO
into the zero air stream will be constant; therefore, a lower flow rate
2
of zero air produces higher concentrations of NO
In the Model T200 the permeation tube enclosure is heated to a constant 50° C (10°
above the maximum operating temperature of the instrument) in order to keep the
permeation rate constant. A thermistor measures the actual temperature and reports it to
the CPU for control feedback.
The flow rate of zero air across the permeation tube is maintained at 50 ± 10 cm³/min
by a critical flow orifice located in the analyzer‟s exhaust manifold.
G
ENERATION
span gas is determined by three factors:
2
(increasing the temperature of the permeation tube
2
Getting Started
)
2
61