Teledyne T200 Operation Manual page 114

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Setup Mode Menus
DISPLAYED AS
= CHANNEL
CONC_OUT_1
=
CONC_OUT_2
=
CONC_OUT_3
=
TEST OUTPUT
=
TEST_OUTPUT replaced
by CONC_OUT_4
(O
Concentration)
2
on analyzers with the
optional O
sensor
2
installed.
These buttons increase /
decrease the analog output
signal level (not the value on the
display)
by 100, 10 or 1 counts.
Continue adjustments until the
voltage measured at the output
of the analyzer and/or the input
of the recording device matches
the value in the upper right hand
corner of the display
(within the tolerances
listed in Table 8.8).
112
Teledyne API – T200 NO/NO
To adjust the signal levels of an analog output channel manually, select the ANALOG
I/O CONFIGURATION submenu (see Figure 5-5) then press:
DIAG
PREV NEXT
DIAG AIO
SET>
A1
Continue pressing SET> until you reach the
A2
A3
A4
DIAG AIO
<SET SET>
DIAG AIO
DIAG AIO
<SET SET>
DIAG AIO
U100 UP10 UP
DIAG AIO
U100 UP10 UP
DIAG AIO
<SET SET> CAL
From the
AIO CONFIGURATION SUBMENU
ANALOG I/O CONFIGURATION
ENTR
AOUTS CALIBRATED: NO
CAL
output to be configured
CONC_OUT_2: 5V, CONC2, NOCAL
EDIT
CONC_OUT_2: RANGE: 5V
EDIT
SET>
Continue pressing SET> until ...
CONC_OUT_2: CALIBRATED:NO
CAL
CONC_OUT_2: VOLT-Z: 0 mV
DOWN DN10 D100 ENTR
CONC_OUT_2: VOLT-S: 4500 mV
DOWN DN10 D100 ENTR EXIT
CONC_OUT_2: CALIBRATED: YES
/NO
Analyzer Operation Manual
2
X
EXIT
EXIT
EXIT
EXIT
EXIT
EXIT
These menus
only appear if
AUTO-CAL is
turned OFF.
EXIT
06858E DCN7057