Advertisement

Teledyne API – T200 NO/NO

5.9.2. ANALOG OUTPUT (DIAG AOUT)

5.9.3. ANALOG I/O CONFIGURATION (DIAG AIO)

Table 5-5: DIAG - Analog I/O Functions
SUB MENU
AOUT CALIBRATED
CONC_OUT_1
CONC_OUT_2
CONC_OUT_3
TEST OUTPUT
AIN CALIBRATED
XIN1
.
.
.
XIN8
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
Use Analog Output to verify functionality and accuracy of the analog outputs. The test
forces all analog output channels to produce signals ranging from 0% to 100% of the full
scale range in 20% increments. This test is useful to verify the operation of the data
logging/recording devices attached to the analyzer.
Section 12.7.6.1 presents instructions for use in troubleshooting and service.
The T200 analyzer comes equipped with four analog outputs. The first three outputs
(A1 A2, & A3) carry analog signals that represent the currently measured concentrating
of NO
, NO and NO
(see Section 5.4.2.1). The fourth output (A4) outputs a signal that
x
2
can be set to represent the current value of one of several test functions (see Table 5-9).
The following table lists the analog I/O functions that are available in the T200 analyzer.
Initiates a calibration of the A1, A2, A3 and A4 analog output channels that
determines the slope and offset inherent in the circuitry of each output.
These values are stored and applied to the output signals by the CPU
automatically.
Sets the basic electronic configuration of the A1 output (NO
There are four options:
 RANGE
1
: Selects the signal type (voltage or current loop) and level of the
output
 REC OFS: Allows them input of a DC offset to let the user manually adjust
the output level
 AUTO CAL: Enables / Disables the AOUT CALIBRATION Feature
 CALIBRATED: Performs the same calibration as AOUT CALIBRATED, but
on this one channel only.
 Same as for CONC_OUT_1 but for analog channel A2 (NO Concentration)
 Same as for CONC_OUT_1 but for analog channel A3 (NO
 Same as for CONC_OUT_1 but for analog channel A4 (TEST CHANNEL)
Initiates a calibration of the A-to-D Converter circuit located on the Motherboard.
For each of 8 external analog inputs channels, shows the gain, offset,
engineering units, and whether the channel is to show up as a Test
function.
FUNCTION
x
Setup Mode Menus
MANUAL
SECTION
5.9.3.1
Concentration).
5.9.2
Concentration)
2
5.9.4
5.9.3.10
5.9.3.11
105