Troubleshooting & Service
12.7.6.4. CONTROL INPUTS
254
Teledyne API – T200 NO/NO2/NOx Analyzer Operation Manual
The control input bits can be tested by applying a trigger voltage to an input and
watching changes in the status of the associated function under the SIGNAL I/O
submenu:
EXAMPLE: to test the "A" control input:
1. Under the DIAG Signal I/O menu (see Section 12.1.3), scroll through the inputs
and outputs until you get to the output named EXT_ZERO_CAL.
2. Connect a jumper from the "+" pin on the appropriate connector to the "U" on the
same connector.
3. Connect a second jumper from the "" pin on the connector to the "A" pin.
4. The status of EXT_ZERO_CAL should change to read "ON".
5. Connect a second jumper from the "" pin on the connector to the "B" pin.
6. The status of EXT_ZERO_CAL should change to read "ON".
Table 12-11: T200 Control Input Pin Assignments and Corresponding Signal I/O
Functions
INPUT
A
B
C, D, E& F
CORRESPONDING I/O SIGNAL
EXT_ZERO_CAL
EXT_SPAN_CAL1
NOT USED
06858E DCN7057