Dc Power Supply - Teledyne T200 Operation Manual

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Teledyne API – T200 NO/NO

12.7.2. DC POWER SUPPLY

Figure 12-4:
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
WARNING – E
Should the AC power circuit breaker trip, investigate and correct the condition
causing this situation before turning the analyzer back on.
If you have determined that the analyzer‟s AC mains power is working, but the unit is
still not operating properly, there may be a problem with one of the instrument‟s
switching power supplies. The supplies can have two faults, namely no DC output, and
noisy output.
To assist tracing DC Power Supply problems, the wiring used to connect the various
printed circuit assemblies and DC Powered components and the associated test points on
the relay PCA follow a standard color-coding scheme as defined in the following table.
Table 12-6:
DC Power Test Point and Wiring Color Codes
NAME
TEST POINT#
DGND
1
+5V
2
AGND
3
+15V
4
-15V
5
+12R
6
+12V
7
TP1
TP2
TP3
TP4
DGND +5V AGND +15V -15V +12R 12V
Location of DC Power Test Points on Relay PCA
A voltmeter should be used to verify that the DC voltages are correct per the values in
the table below, and an oscilloscope, in AC mode, with band limiting turned on, can be
used to evaluate if the supplies are producing excessive noise (> 100 mV p-p).
LECTRICAL
COLOR
DEFINITION
Black
Digital ground
Red
Green
Analog ground
Blue
Yellow
Purple
(ground) line
Orange
TP5
TP6
TP7
Troubleshooting & Service
S
H
HOCK
AZARD
12 V return
249