Setup Mode Menus
DISPLAYED AS
= CHANNEL
CONC_OUT_1
=
A1
CONC_OUT_2
=
A2
CONC_OUT_3
=
A3
TEST OUTPUT
=
A4
TEST_OUTPUT replaced
by CONC_OUT_4
(O
Concentration)
2
on analyzers with the
optional O
sensor
2
installed.
Analyzer automatically
calibrates the DCV signal
output from the analog
output channel to the
VDC-to-mA converter.
These button increase /
decrease the analog output
signal level (not the value on the
display)
by 100, 10 or 1 counts.
Continue adjustments until the
current signal measured at the
output of the analyzer matches
the zero and span points of the
intended current range
(e.g. 0 mA–20 mA;
4 mA–20 mA)
114
Teledyne API – T200 NO/NO
To adjust the zero and span signal levels of the current outputs, select the ANALOG I/O
CONFIGURATION submenu (see Figure 5-5) then press:
AIO CONFIGURATION SUBMENU
DIAG
ANALOG I/O CONFIGURATION
PREV NEXT
DIAG AIO
SET>
Continue pressing SET> until you reach the
DIAG AIO
CONC_OUT_2: 5V, CONC2, NOCAL
<SET SET>
DIAG AIO
SET>
Continue pressing SET> until ...
DIAG AIO
<SET SET>
DIAG AIO
AUTO CALIBRATING CONC_OUT_2
DIAG AIO
U100 UP10 UP
DIAG AIO
U100 UP10 UP
DIAG AIO
<SET SET> CAL
From the
ENTR
AOUTS CALIBRATED: NO
CAL
output to be configured
EDIT
CONC_OUT_2: RANGE: CURR
EDIT
CONC_OUT_2: CALIBRATED:NO
CAL
CONC_OUT_2: CURR-Z: 0 mV
DOWN DN10 D100 ENTR
CONC_OUT_2: CURR-S: 5000 mV
DOWN DN10 D100 ENTR EXIT
CONC_OUT_2: CALIBRATED: YES
/NO
Analyzer Operation Manual
2
X
EXIT
EXIT
EXIT
EXIT
EXIT
EXIT
These menus
adjust the mAmp
signal output by
converter circuit.
EXIT