9.1.3. DATA RECORDING DEVICES

9.1.4. NO

CONVERSION EFFICIENCY (CE)

2
9.2. MANUAL CALIBRATION CHECKS AND CALIBRATION OF
THE T200 ANALYZER IN ITS BASE CONFIGURATION
I
MPORTANT
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
A strip chart recorder, data acquisition system or digital data acquisition system should
be used to record data from the serial or analog outputs of the T200.
If analog readings are used, the response of the recording system should be
checked against a NIST traceable voltage source or meter.
Data recording devices should be capable of bi-polar operation so that negative
readings can be recorded.
For electronic data recording, the T200 provides an internal data acquisition system
(DAS), which is described in detail in Section 7.
APICOM, a remote control program, is also provided as a convenient and powerful tool
for data handling, download, storage, quick check and plotting (see Sections 7.2.1, and
the APICOM software manual downloadable from:
In order for the NO
converter to function properly, oxygen must be present in the
2
sample stream. In addition, to ensure accurate operation of the T200, it is important to
check the NO
conversion efficiency (CE) periodically and to update this value as
2
necessary.
See Section 12.7.10 for instructions on checking or calculating the current NO
NO converter efficiency using T200's onboard firmware.
See Section 12.7.11 for instructions on checking or calculating the current NO
NO converter efficiency using a simplified Gas Phase Titration Method.
ZERO/SPAN CALIBRATION CHECKS VS. ZERO/SPAN CALIBRATION
Pressing the ENTR button during the following procedure resets the stored
values for OFFSET and SLOPE and alters the instrument's Calibration.
This should ONLY BE DONE during an actual calibration of the T200.
NEVER press the ENTR button if you are only checking calibration.