9.2.1. SETUP FOR BASIC CALIBRATION CHECKS AND CALIBRATION

Set up for Manual Calibrations/Checks of T200's in Base Configuration w/ a Gas Dilution
Figure 9-1:
Calibrator
Set up for Manual Calibrations/Checks of T200's in Base Configuration w/ Bottled Gas
Figure 9-2:
182
Teledyne API – T200 NO/NO
Connect the sources of zero air and span gas as shown below in one of the following
ways:
VENT
MODEL 701
Zero Gas
Generator
Vent here if output of calibrator
is not already vented.
VENT
here if input
is pressurized
3-way Valve
Manual
Control Valve
here if input
is pressurized
MODEL 700E
Gas Dilution
Calibrator
PUMP
MODEL 701
Zero Gas
Generator
PUMP
/NO
Analyzer Operation Manual
2
X
Source of
SAMPLE GAS
Removed during
calibration
SAMPLE
EXHAUST
Chassis
Source of
SAMPLE GAS
Removed during
calibration
SAMPLE
EXHAUST
Chassis
06858E DCN7057