Teledyne API – T200 NO/NO
The NO
& NO span concentration
X
values automatically default to
400.0 PPB
If this is not the the concentration of
the span gas being used, toggle
these buttons to set the correct
concentration of the NO
and NO
X
calibration gases.
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
To verify that the analyzer span setting is set for 400 PPB, press:
SAMPLE
<TST
TST>
CAL
M-P CAL
RANGE=500.0 PPB
<TST TST> ZERO SPAN
M-P CAL
CONCENTRATION MENU
CONV
NOX
NO
M-P CAL
[GAS TYPE] SPAN CONC:400.0 Conc
0
4
I
Z
/S
C
NITIAL
ERO
PAN
To perform an initial calibration, press:
RANGE=500.0 PPB
NOX= XXXX
NOX=XXXX
CONC
0
0
.0
ENTR EXIT
P
ALIBRATION
ROCEDURE
Getting Started
SETUP
EXIT
EXIT
EXIT ignores the new
setting and returns to
the previous display.
ENTR accepts the new
setting and returns to the
CONCENTRATION MENU.
If using NO span gas in
addition to NO
repeat last
X
step.
71