Changing The Odryer Particulate Filter - Teledyne T200 Operation Manual

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Teledyne API – T200 NO/NO
11.3.2. CHANGING THE O
ATTENTION
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
DRYER PARTICULATE FILTER
3
The air for the O
generator passes through a dryer equipped with a small particulate
3
filter at its inlet, which prevents dust from entering the ozone dryer and degrading the
dryer‟s performance over time. Change the filter according to the service interval in
Table 11-1 as follows:
1. Before starting the procedure, check and record the average RCEL pressure and
the OZONE FLOW values.
2. Turn off the analyzer, unplug the power cord and remove the cover.
3. Unscrew the nut around the port of the filter using two 5/8" wrenches.
COULD DAMAGE INSTRUMENT AND VOID WARRANTY
Ensure to use proper wrenches.
Hold the main dryer fitting with a 5/8‖ wrench to ensure that it does not turn
against the dryer.
Performing this procedure improperly or with incorrect tools creates a risk of
causing a significant leak.
4. Take off the old filter element and replace it with a suitable equivalent (Teledyne
API P/N FL-3).
Figure 11-2:
Particle Filter on O
5. Hold the main dryer fitting steady with a 5/8" wrench and tighten the nut with your
hands.
If necessary use a second wrench but do not over-tighten the nut.
6. Replace the cover, plug in the power cord and restart the analyzer.
Instrument Maintenance
Supply Air Dryer
3
211