Table 9-4: Example Autocal Sequence - Teledyne T200 Operation Manual

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Teledyne API – T200 NO/NO
I
MPORTANT
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
The following example sets sequence #2 to do a zero-span calibration every other day
starting at 1:00 AM on September 4, 2011, lasting 15 minutes, without calibration. This
will start ½ hour later each iteration.

Table 9-4: Example AutoCal Sequence

MODE AND
ATTRIBUTE
SEQUENCE
MODE
TIMER ENABLE
STARTING DATE
STARTING TIME
DELTA DAYS
DELTA TIME
DURATION
CALIBRATE
The programmed STARTING_TIME must be a minimum of 5 minutes later than the
real time clock for setting real time clock (See Section 5.6).
Avoid setting two or more sequences at the same time of the day.
Any new sequence that is initiated whether from a timer, the COM ports or the
contact closure inputs will override any sequence that is in progress.
The CALIBRATE attribute must always be set to OFF on analyzers with IZS Options
installed and functioning.
Calibrations should ONLY be performed using external sources of Zero Air and
Span Gas whose accuracy is traceable to EPA standards.
VALUE
2
ZERO-SPAN
ON
Sept. 4, 2011
1:00 AM
2
00:30
15.0
OFF
I
R
MPACT ON
EADINGS OR
Calibration Procedures
COMMENT
Define Sequence #2
Select Zero and
Span Mode
Enable the timer
Start after
Sept 4, 2011
First Span starts at
1:00AM
Do Sequence #2
every other day
Do Sequence #2 ½
hr later each day
Operate Span valve
for 15 min
Calibrate at end of
Sequence
D
ATA
199

Advertisement

Table of Contents
loading

This manual is also suitable for:

T200uT201T265T200u-noyT200up

Table of Contents