Maintaining The External Sample Pump (Pump Pack) - Teledyne T200 Operation Manual

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Instrument Maintenance

11.3.4. MAINTAINING THE EXTERNAL SAMPLE PUMP (PUMP PACK)

11.3.4.1. REBUILDING THE PUMP
11.3.4.2. REPLACING THE SCRUBBER
214
Teledyne API – T200 NO/NO
Do not over-tighten the fittings.
15. If the service interval for this item has been exceeded, it may also be necessary to
clean the reaction cell as described in Section 11.3.9.
16. Leak check the system using the pressurized approach described in Section
11.3.12.2.
If necessary, tighten the fittings some more but do not over-tighten.
17. Restart the analyzer and pump and continue operation.
18. Recalibrate the analyzer after one hour (Section 9).
If Auto Zero is high or is changing/not constant, you may have to wait a day until
the silica gel is conditioned before recalibrating the instrument.
The sample pump head periodically wears out and must be replaced when the RCEL
pressure exceeds 10 in-Hg-A (at sea level, adjust this value accordingly for elevated
locations).
A pump rebuild kit is available from the factory. Refer to the label on the pump for
the part number. Instructions and diagrams are included in the kit.
A flow and leak check after rebuilding the sample pump is recommended.
A span check and re-calibration after this procedure is necessary as the response
of the analyzer changes with the RCEL pressure.
Do NOT attempt to change the contents of the inline exhaust scrubber
cartridge; change the entire cartridge
1. Through the SETUP>MORE>DIAG menu turn OFF the OZONE GEN OVERRIDE.
Wait 10 minutes to allow pump to pull room air through scrubber before proceeding
to step 2.
2. Disconnect exhaust line from analyzer.
3. Turn off (unplug) analyzer sample pump.
4. Disconnect tubing from (NOx or charcoal) scrubber cartridge.
5. Remove scrubber from system.
6. Dispose of according to local laws.
7. Install new scrubber into system.
8. Reconnect tubing to scrubber and analyzer.
9. Turn on pump.
Through the SETUP menu (per Step 1 above) turn ON the OZONE GEN
10.
OVERRIDE
.
/NO
Analyzer Operation Manual
2
X
CAUTION!
.
06858E DCN7057