Teledyne T200 Operation Manual page 166

No/no2/nox analyzer
Hide thumbs Also See for T200:
Table of Contents

Advertisement

Data Acquisition System (DAS) and APICOM
7.1.3.8. HOLDOFF FEATURE
Use the PREV and
NEXT buttons to
scroll to the DATA
CHANNEL to be
edited.
Toggle these buttons
to turn the HOLDOFF
feature ON/OFF.
164
Teledyne API – T200 NO/NO
The DAS HOLDOFF feature prevents data collection during calibration operations.
To enable or disable the HOLDOFF, follow the instruction shown in Section 7.1.3 then
press:
Starting at the EDIT CHANNEL MENU
SETUP X.X
0) CONC: ATIMER 5, 800
PREV NEXT
INS
SETUP X.X
NAME: CONC
EDIT
<SET
SET>
Continue pressing <SET or SET> until ...
SETUP X.X
CAL.HOLD OFF: OFF
<SET
SET> EDIT
SETUP X.X
CAL.HOLD OFF: OFF
OFF
HOLDOFF also prevents DAS measurements from being made at certain times when
the quality of the analyzer‟s sample measurements may be suspect (e.g. while the
instrument is warming up). In this case, the length of time that the HOLDOFF feature
is active is determined by the value of the internal variable (VARS), DAS_HOLDOFF.
To set the length of the DAS_HOLDOFF period, go to the SETUP>MORE>VARS
menu and at the DAS_HOLDOFF parameter (see Table 5-3), press the Edit button.
/NO
2
DEL
PRNT
EDIT
EXIT
EXIT
EXIT
ENTR EXIT
Analyzer Operation Manual
X
EXIT discards the new
setting.
ENTR accepts the
new setting.
06858E DCN7057