5.9.3.9. ADDING A RECORDER OFFSET TO AN ANALOG OUTPUT
Toggle these
buttons to set
the desired
offset value.
06858E DCN7057
/NO
Analyzer Operation Manual
2
X
Some analog signal recorders require that the zero signal be significantly different from
the baseline of the recorder in order to record slightly negative readings from noise
around the zero point. This can be achieved in the T200 by defining a zero offset, a
small voltage (e.g., 10% of span).
To add a zero offset to a specific analog output channel, select the ANALOG I/O
CONFIGURATION submenu (see Figure 5-5) then press:
DIAG
PREV NEXT
DIAG AIO
SET>
Continue pressing SET> until you reach the
DIAG AIO
<SET SET>
DIAG AIO
SET>
DIAG AIO
<SET SET>
DIAG AIO
+
DIAG AIO
EXAMPLE
DIAG AIO
<SET SET> EDIT
From the
AIO CONFIGURATION SUBMENU
ANALOG I/O CONFIGURATION
ENTR
AOUTS CALIBRATED: NO
CAL
output to be configured
CONC_OUT_2: 5V, OVR, NOCAL
EDIT
CONC_OUT_2 RANGE: 5V
EDIT
Continue pressing SET> until ...
CONC_OUT_2: REC OFS: 0 mV
EDIT
CONC_OUT_2: REC OFS: 0 mV
0
0
0
0
CONC_OUT_2: REC OFS: -10 mV
0
0
0
1
CONC_OUT_2: REC OFS: -10 mV
Setup Mode Menus
EXIT
EXIT
EXIT
EXIT
EXIT
ENTR EXIT
EXIT
ENTR
EXIT
117