Troubleshooting & Service
12.7.12.2. ELECTRICAL TEST
SAMPLE
RANGE=500.0 PPB
<TST
TST>
CAL
Continue pressing <TST or TST> until ...
SAMPLE
PMT=2750 MV
<TST
TST> CAL
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK
SETUP X.X
SECONDARY SETUP MENU
COMM VARS
DIAG
266
Teledyne API – T200 NO/NO2/NOx Analyzer Operation Manual
The electrical test function creates a current, which is substituted for the PMT signal and
feeds it into the preamplifier board.
This signal is generated by circuitry on the pre-amplifier board itself and tests the
filtering and amplification functions of that assembly along with the A/D converter on
the motherboard.
It does not test the PMT itself.
The electrical test should produce a PMT signal of about 2000 ±1000 mV.
To activate the electrical test, press:
NOX= XXXX
SETUP
NOX= XXXX
SETUP
MORE
EXIT
EXIT
While the ETEST is
active PMT should =
2000 mv ± 1000mv
Concentration display
continuously cycles
through all gasses.
SETUP X.X
ENTER PASSWORD
8
1
8
DIAG
SIGNAL I/O
PREV
NEXT
Continue pressing NEXT until ...
DIAG OPTIC
ELECTRICAL TEST
PREV NEXT
DIAG ELEC
PMT=2750 MV
<TST
TST> CAL
ENTR
EXIT
ENTR
EXIT
ENTR
EXIT
NOX= XXXX
EXIT
06858E DCN7057