Instrument Maintenance
220
Teledyne API – T200 NO/NO
Figure 11-5:
10. Wrap the band heater around the new replacement converter cartridge and tighten
the screws using a high-temperature anti-seize agent (Teledyne API P/N CH42)
such as copper paste.
Ensure to use proper alignment of the heater with respect to the converter
tubes.
11. Replace the Moly converter assembly by routing the cables through the holes in the
converter housing and reconnecting them properly.
12. Reconnect the steel cable clamp around the power leads for safe operation.
13. Reattach the tube fittings to the converter and replace the Moly insulation (top layer
and corner cut out layers).
14. Reinstall the converter assembly cover.
15. Reinstall the instrument cover and power up the analyzer.
16. Allow the converter to burn-in for 24 hours, and then recalibrate the instrument.