Figure 4-25: Analog Output - Voltage Adjustment - Teledyne T100 Instruction Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for T100:
Table of Contents

Advertisement

Operating Instructions
DIAG
ANALOG I / O CONFIGURATION
PREV
NEXT
DIAG AIO
AOUTS CALIBRATED: NO
< SET
SET>
CAL
Press SET> to select the analog output channel to be configured:
DISPLAYED AS =
CONC_OUT_1 =
CONC_OUT_2 =
TEST OUTPUT =
DIAG AIO
CONC_OUT_1 :5V, NO CAL
< SET SET>
EDIT
DIAG AIO
CONC_OUT_1 RANGE: 5V
SET>
EDIT
These keys increase / decrease the analog
output by 100, 10 or 1 counts.
Continue adjustments until the voltage measured
at the output of the analyzer and/or the input of
the recording device matches the value in the
upper right hand corner of the display to the
tolerance listed in Table 6-10.
The concentration display will not change. Only
the voltage reading of your voltmeter will change.
112
To make these manual adjustments, the AOUT auto-calibration feature must be turned
OFF
(refer
to
CONFIGURATION MENU from the DIAG Menu (refer to Figure 4-17), then press:
ENTR
EXIT
EXIT
CHANNEL
A1
A2
A4
EXIT
EXIT
Analog Output – Voltage Adjustment
Figure 4-25:
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
Section
4.8.4.2).
DIAG AIO
CONC_OUT_1 OVERRANGE: ON
SET>
EDIT
DIAG AIO
CONC_OUT_1 REC OFS: 0 mV
< SET
SET>
EDIT
DIAG AIO
CONC_OUT_1 AUTO CAL: OFF
< SET
SET>
EDIT
DIAG AIO
CONC_OUT_1 CALIBRATED: NO
< SET
CAL
DIAG AIO
CONC_OUT_1 VOLT–Z : 0 mV
U100 UP10 UP DOWN DN10 D100 ENTR EXIT
DIAG AIO
CONC_OUT_1 VOLT–S : 4500 mV
U100 UP10 UP DOWN DN10 D100 ENTR EXIT
DIAG AIO
CONC_OUT_1 CALIBRATED: YES
< SET
CAL
Activate
the
ANALOG
EXIT
EXIT
If AutoCal is ON, go to
Section 6.7.3
EXIT
EXIT
EXIT ignores the
new setting.
ENTR accepts the
new setting.
EXIT
I/O