Teledyne T400 User Manual

Teledyne T400 User Manual

Photometric ozone analyzer with numaview software
Table of Contents

Advertisement

Quick Links

Photometric Ozone Analyzer
2019
Copyright
API
Teledyne
User Manual
Model T400
with NumaView™ software
© Teledyne API (TAPI)
9970 Carroll Canyon Road
San Diego, CA 92131-1106
USA
Toll-free Phone: +1 800-324-5190
Phone: +1 858-657-9800
Fax: +1 858-657-9816
Email:
api-
sales
Website:
http://
@teledyne.com
.teledyne-api.com/
www
083730400A DCN8130
31 May 2019

Advertisement

Table of Contents
loading

Summary of Contents for Teledyne T400

  • Page 1 User Manual Model T400 Photometric Ozone Analyzer with NumaView™ software © Teledyne API (TAPI) 9970 Carroll Canyon Road San Diego, CA 92131-1106 Toll-free Phone: +1 800-324-5190 Phone: +1 858-657-9800 Fax: +1 858-657-9816 Email: api- @teledyne.com sales Website: http:// .teledyne-api.com/ 2019...
  • Page 3 NOTICE OF COPYRIGHT © 2019 Teledyne API (TAPI). All rights reserved. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 4 TRADEMARKS All trademarks, registered trademarks, brand names or product names appearing in this document are the property of their respective owners and are used herein for identification purposes only. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 5 NEVER use any gas analyzer to sample combustible gas(es)! For Technical Assistance regarding the use and maintenance of this instrument or any other Teledyne API product, contact Teledyne API’s Technical Support Department: Telephone: +1 800-324-5190 (toll free) or +1 858-657-9800 Email: api-techsupport@teledyne.com...
  • Page 6 Si vous utilisez cet instrument d’une autre manière que celle pour laquelle il a été prévu, l’instrument pourrait se comporter de façon imprévisible et entraîner des conséquences dangereuses. NE JAMAIS utiliser un analyseur de gaz pour échantillonner des gaz combustibles! Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 7 PRODUCT RETURN All units or components returned to Teledyne API should be properly packed for handling and returned freight prepaid to the nearest designated Service Center. After the repair, the equipment will be returned, freight prepaid.
  • Page 8 ABOUT THIS MANUAL This manual is comprised of multiple documents, in PDF format, as listed below. Part No. Name/Description 083730400 T400 Photometric Ozone Analyzer User Manual with NumaView™ Software (this manual) MODBUS Registers, Appendix A 069130000 T400 Interconnect Diagram, Appendix B...
  • Page 9 Provides information about that which could either affect accuracy of instrument readings or cause loss of data. Provides information pertinent to the proper care, NOTE operation or maintenance of the instrument or its parts. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 10: Table Of Contents

    2.5.1.2 Configuring Trigger Types: Conditional ................63 2.5.1.3 Downloading DAS (Data Acquisition System) Data ............63 2.5.2 Setup>Events ...........................64 2.5.2.1 Editing or Deleting Events ....................66 2.5.2.2 Using Events as Triggers for Data Logging...............67 viii Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 11 4.6 EPA Protocol Calibration ..........................96 5. Maintenance and Service ......................97 5.1 Maintenance Schedule ..........................97 5.2 Predicting Diagnostics ..........................100 5.3 Operational Health Checks ........................101 5.4 Software/Firmware Updates ........................101 5.4.1 Remote Updates ..........................101 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 12 5.7.10 Motherboard ...........................129 5.7.10.1 A/D Functions ........................129 5.7.10.2 Status Outputs .........................129 5.7.10.3 Control Inputs – Remote Zero, Span ................130 5.7.11 CPU ..............................130 5.7.12 RS-232 Communications .......................130 5.7.12.1 General RS-232 Troubleshooting ..................130 Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 13 6.3.5.2 AC Configuration – Heaters for Option Packages (JP6) ..........158 6.3.6 Photometer Layout and Operation ....................159 6.3.7 Front Panel Touchscreen/Display Interface ...................160 6.1.1.1 LVDS Transmitter Board ....................161 6.1.1.2 Front Panel Touchscreen/Display Interface PCA ............161 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 14 6.2 Software Operation ............................161 6.2.1 Adaptive Filter ..........................162 6.2.2 Temperature/Pressure Compensation (TPC) ................162 6.2.3 Calibration - Slope and Offset ......................163 Appendix A – MODBUS Registers Appendix B – Interconnect Diagram Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 15 Figure 2-15. Gas Line Connections for the T400 Analyzer with IZS Option (OPT 50G) ..........43 Figure 2-16. Gas Line Connections when the T400 Analyzer is Located in the Room Being Monitored ....44 Figure 2-17. Gas Line Connections when the T400 Analyzer is Monitoring a Remote Location ......45 Figure 2-18.
  • Page 16 Figure 6-15. Typical Jumper Set (JP2) Set Up of Optional Metal Wool Scrubber Heater ........159 Figure 6-16. O Photometer Layout – Top Cover Removed ..................160 Figure 6-17. Front Panel and Display Interface Block Diagram ................161 Figure 6-18. Basic Software Operation ........................162 Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 17 Table 2-13. COM1/COM2 Configuration ........................76 Table 2-14. LAN/Ethernet Configuration Properties ....................78 Table 3-1. Ethernet Status Indicators ........................81 Table 3-2. Teledyne API's Hessen Protocol Response Modes ................85 Table 3-3. Default Hessen Status Flag Assignments ....................86 Table 4-1. AUTO CAL States ...........................93 Table 4-2.
  • Page 18: Introduction, Specifications, Approvals, & Compliance

    In the T400 the intensity of an ultra violet light is measured after it passes through a chamber, called the sample cell, where it is absorbed in proportion to the amount of ozone present.
  • Page 19: Specifications

    SPECIFICATIONS Table 1-1. T400 Basic Unit Specifications PARAMETER SPECIFICATION Ranges Min: 0-100 ppb Full scale Max: 0-10 ppm Full scale (selectable, dual ranges and auto-ranging supported) Measurement Units ppb, ppm, µg/m , mg/m (selectable) Zero Noise < 0.2 ppb (RMS) (with 80 Sample Digital Filter) Span Noise <...
  • Page 20: Epa Designation

    < 5 min to 95% EPA DESIGNATION Teledyne API’s Model T400 photometric ozone analyzer is officially designated as US EPA Federal Equivalent Method (FEM), Designation Number EQOA-0992-087. The official “List of Designated Reference and Equivalent Methods” is published in the U.S.
  • Page 21: Getting Started

    Do not operate this instrument without first removing dust plugs from SAMPLE and EXHAUST ports on the rear panel. Note Teledyne API recommends that you store shipping containers and materials for future use if/when the instrument should be returned to the factory for repair and/or calibration service. See Warranty statement in this manual and Return Merchandise Authorization (RMA) on our Website at http://www.teledyne-api.com.
  • Page 22: Ventilation Clearance

    Sides of the instrument 2.5 cm / 1 in 2.5 cm / 1 in Above and below the instrument Various rack mount kits are available for this analyzer. Contact Sales for more information. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 23: Instrument Layout

    The front panel (Figure 2-1) includes two USB ports for peripheral device connections, which can be used with mouse and keyboard as alternatives to the touchscreen interface, or with flash drive for uploads/downloads (devices not included). Figure 2-1. Front Panel Layout 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 24: Rear Panel

    2.2.2 R EAR PANEL Figure 2-2 shows the layout of the rear panel. Figure 2-2. Rear Panel Layout Table 2-2 provides a description of each component on the rear panel. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 25: Table 2-2. Rear Panel Description

    Option for concentration alarms and system warnings. ETHERNET Connector for network or Internet remote communication, using Ethernet cable. (option) USB Connector for direct connection to laptop computer, using USB cable. Information Label Includes voltage and frequency specifications 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 26: Internal Chassis

    HASSIS CAUTION – UV RADIATION RISK Do not look directly at the light of the UV lamp. Use proper UV-protective lenses. Figure 2-3. T400 Internal Layout – Top View with IZS Option Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 27: Connections And Startup

    AC voltage range and ensure that it is equipped with a functioning earth ground. It is important to adhere to all safety and cautionary messages. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 28: Connecting Analog Outputs

    Following Figure 2-5 are instructions for converting current loop analog outputs to standard 0-to-5 VDC outputs. To calibrate or adjust these outputs use the Setup>Analog Outputs menu (Section Error! Reference source not found.). Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 29: Figure 2-5. Current Loop Option Installed On Motherboard

    For information on preventing ESD damage, refer to the manual, Fundamentals of ESD, PN 04786, which can be downloaded from our website at http://www.teledyne-api.com. Figure 2-5. Current Loop Option Installed on Motherboard 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 30 7. Reattach the top case to the analyzer. • The analyzer is now ready to have a voltage-sensing, recording device attached to that output. 8. Calibrate the analog output in the Setup>Analog Outputs>Analog Output Cal menu. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 31: Connecting The Status Outputs (Digital Outputs)

    + 5 VDC, 300 mA source maximum The ground level from the analyzer’s internal DC power Digital Ground supplies. This connection should be used as the ground return when +5VDC power is used. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 32: Connecting The Control Inputs (Digital Inputs)

    Internally generated 5V DC power. To activate inputs A – F, place a jumper 5 VDC output between this pin and the “U” pin. The maximum amperage through this port is 300 mA (combined with the analog output supply, if used). Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 33: Connecting The Concentration Alarm Relay (Option 61)

    Connect a USB cable between the analyzer and computer USB ports. The baud rate for the analyzer and the computer must match; see Section 3.2. If this option is installed, the COM2 port cannot be used for Note anything other than Multidrop communication. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 34: Figure 2-9. Rear Panel Connector Pin-Outs For Rs-232 Mode

    Check cables acquired from sources not from Teledyne API for pin assignments (Figure 2-9) before using. Electronically, the difference between the DCE and DTE is the pin assignment of the Data Receive and Data Transmit functions.
  • Page 35: Figure 2-10. Default Pin Assignments For Cpu Com Port Connector (Rs-232)

    • Parity: None COM2: RS-232 (configurable to RS-485), DB-9 female connector. • • Baud rate: 19200 bits per second (baud) • Data Bits: 8 data bits with 1 stop bit • Parity: None 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 36 (Note that the CPU’s COM2 connector is not used in Multidrop) J4 on the Multidrop/LVDS PCA to J12 on the motherboard • J1 on the Multidrop/LVDS PCS to the front panel LCD • Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 37: Figure 2-11. Jumper And Cables For Multidrop Mode

    (Ensure you are using the correct RS-232 cables internally wired specifically for RS- 232 communication; see Error! Reference source not found., “Communication Cables” and Section 2.3.1.7: Connecting the Communications Inerfaces, “RS-232 Connection”). 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 38: Figure 2-12. Rs-232-Multidrop Pca Host/Analyzer Interconnect Diagram

    Edit button to set COM1 Quiet Mode and COM1 Multidrop to ENABLED; then press the Accept button. 9. Also check the COM1 Baud Rate to ensure it reads the same for all instruments (edit if needed and press the Accept button). Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 39: Pneumatic Connections

    The (communication) Host instrument can only • address one instrument at a time. Teledyne API recommends setting up the first link, • between the Host and the first analyzer, and testing it before setting up the rest of the chain.
  • Page 40: Important Information About Calibration Gases

    Internal Zero Span (IZS) or an external zero air scrubber option, it is capable of creating zero air. For analyzers without an IZS or external zero air scrubber option, a zero air generator such as the Teledyne API Model T701 can be used. Teledyne API – Model T400 O...
  • Page 41: Pneumatic Connections: Basic Configuration

    (O ), this calibration gas cannot be supplied in precisely calibrated bottles like other gases. If the T400 analyzer is not equipped with the optional internal zero air generator (IZS), an • external O generator capable of supplying accurate O calibration mixtures must be used.
  • Page 42 For the Model T400 photometric ozone analyzer in its basic configuration (i.e. without the optional internal zero air source or valves), attach the following pneumatic lines: AMPLE OURCE Attach a sample inlet line to the SAMPLE inlet port. Sample Gas pressure must equal ambient atmospheric pressure (1.0 psig) •...
  • Page 43: Pneumatic Connections: Zero/Span (Z/S) Option

    Restrictor to regulate flow at 2 x’s analyzer gas flow Figure 2-14. Gas Line Connections for the T400 Analyzer with Zero/Span Valve Option (OPT 50A) The instrument’s zero air and span gas flow rate required for this option is 800 cc/min, however, the US EPA recommends that the cal gas flow rate be at least 1600 cc/min.
  • Page 44 Attach a gas line from the source of zero air (e.g. a Teledyne API M701 zero air Generator) to the ZERO AIR inlet. • Zero air can be supplied by the API M701 zero air generator. • A restrictor is required to regulate the gas flow at 2x the gas flow of the analyzer.
  • Page 45: Pneumatic Connections: Internal Zero/Span (Izs) Option

    • Vented outside the shelter or immediate area surrounding the instrument OURCE Attach a gas line from the source of zero air (e.g., a Teledyne API M701 zero air Generator) to the DRY AIR inlet. The gas from this line will be used internally as zero air and as source air for the internal •...
  • Page 46: Pneumatic Connections: Ambient Air Monitoring In Same Room

    ZERO AIR DRY AIR SAMPLE and DRY AIR inlets open to room Figure 2-16. Gas Line Connections when the T400 Analyzer is Located in the Room Being Monitored AMPLE AS AND OURCES For instruments located in the same room being monitored, there is no need to attach the gas inlet lines to the SAMPLE inlet or the DRY AIR inlet.
  • Page 47: Pneumatic Connections: Ambient Air Monitoring In Remote Locations

    (with IZS installed) BEING ZERO AIR MONITORED DRY AIR Figure 2-17. Gas Line Connections when the T400 Analyzer is Monitoring a Remote Location AMPLE OURCE Attach a sample inlet line leading from the room being monitored to the SAMPLE inlet fitting.
  • Page 48: Pneumatic Flow Diagrams

    DRY AIR INLET Flow / Pressure SAMPL E Sensor PCA PRESS URE SENSO R EXHAUST Sample Gas OUTLET PUMP Flow Control FLOW SENSO R Figure 2-18. T400 Pneumatic Diagram – Basic Unit Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 49: Pneumatic Flow: Zero/Span (Zs) Valve Option

    Sensor PCA SENSOR EXHAUST Sample Gas OUTLET PUMP Flow Control FLOW SENSOR Figure 2-19. T400 Pneumatic Diagram with Zero/Span (ZS) Option (OPT50A) Table 2-6. Zero/Span Valve Operating States for Option 50A Mode Valve Condition Sample/Cal Open to SAMPLE inlet SAMPLE Zero/Span...
  • Page 50: Pneumatic Flow: Internal Zero/Span (Izs) Valve Option

    PRESSURE Sensor PCA SENSOR EXHAUST OUTLET PUMP FLOW SENSOR Figure 2-20. T400 Pneumatic Diagram with Internal Zero/Span (IZS) Option (OPT50G) Table 2-7. Internal Zero/Span Valve Operating States for Option 50G Mode Valve Condition Sample/Cal Valve Open to SAMPLE inlet SAMPLE...
  • Page 51: Startup, Functional Checks, And Initial Calibration

    Upon initial startup, a sequence of status screens (Figure 2-21) appear prior to the Home page (Figure 2-22). Figure 2-21. Status Screens at Startup Upon any startup, this instrument should warm up for approximately one hour before reliable measurements can be taken. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 52: Alerts: Warnings And Other Messages

    Figure 2-23. Viewing Active Alerts Page If Alerts about warning conditions persist after the warm up period or after being cleared, investigate their cause using the troubleshooting guidelines in Section 5.7. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 53: Functional Checks

    The method for performing a calibration differs slightly depending on whether or not any of the available internal zero air or valve options are installed. Follow the appropriate calibration instructions in Section 4. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 54: Menu Overview

    Instrument Section 2.5.9 connected to a network that is connected to the Internet. Time Zone change requires special procedures (Section 5.5). Comm View and configure network and serial communications. Section 2.5.10 Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 55: Home Page

    Figure 2-25 presents an orientation to the main display screen; Figure 2-26 shows that pressing the gas name or its concentration value or a meter below displays a live plot of their respective readings. Section 2.5.6 provides configuration instructions. Figure 2-25. User Interface Orientation 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 56 Figure 2-26. Concentration and Stability Graph (top) and Meter Graph (bottom) Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 57: Dashboard

    Alerts menu on the Home screen or by pressing the Alerts shortcut (Figure 2-28). While Alerts can be cleared from the Active Alerts page, they remain recorded in the Utilities>Alerts Log menu. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 58 (Figure 2-29). However, Alerts can reappear if the conditions causing them are not resolved. For troubleshooting guidance, refer to Section 5.7. Figure 2-29. Active Alerts Cleared Alerts and Events remain recorded in the Utilities>Alerts Log (Figure 2-30). Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 59: Calibration

    The Calibration menu is used for multipoint calibrations as well as calibrations for zero and span with valve options, and for external calibration with valve options installed. Calibration procedures are presented in Section 4. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 60: Utilities

    Use to transfer instrument configuration from/to other same-model instruments (Section 2.6) 2.4.6 S ETUP The Setup menu is for configuring the instrument’s various programmable features and functions. Section 2.5 provides details for the menus under Setup. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 61: Setup Menu: Features/Functions Configuration

    Figure 2-33 provides illustrated instructions for setting up a new log, with Sections 2.5.1.1 and 2.5.1.2 providing additional details. To transfer captured instrument data to a flash drive see Section 2.5.1.3. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 62 Figure 2-31. Datalog Configuration, New Log Page Figure 2-32. Datalog Configuration, Existing Log Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 63 Figure 2-33. Creating a New Data Log The parameters available in the list of Log Tags include the names of Events configured in the Events page (Section 2.5.2). 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 64: Configuring Trigger Types: Periodic

    Periodic Trigger requires an interval that is set to number of minutes and a start time that is set to date and clock time. Figure 2-34. Datalog Periodic Trigger Configuration Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 65: Configuring Trigger Types: Conditional

    2. Select all or define a period from which to download the collected data. 3. Press the Download button, and when complete, as indicated in the Status field, press the Done button (changed from “Cancel”) and remove the flash drive. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 66: Setup>Events

    Home>Setup>Events menu (Figure 2-37). Press ADD to create a new Event (refer to Figure 2-38 for details), or select an existing Event to either Edit or Delete it (Figure 2-40). Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 67 Non-latching allows the entry in the Active Alerts screen and the internal Event state to continuously update based on the Event criteria, requiring no user interaction to clear the Alert or Event state). 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 68: Editing Or Deleting Events

    Select an Event from the list (Figure 2-37) and press the Edit button to view or edit the details (Figure 2-39), or press the Delete button to delete the Event. Figure 2-40. Edit or Delete an Event Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 69: Using Events As Triggers For Data Logging

    Name field of the Events Configuration page will appear in the list of Log Tags of the Datalog Configuration page. The Data Logger is presented in Section 2.5.1. 2.5.3 S >D ETUP ASHBOARD Figure 2-41. Dashboard Display and Configuration 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 70: Setup>Autocal (With Valve Option)

    Controls range mode, single (SNGL) or dual (DUAL). System Hours Total system runtime hours TPC Enable Enables or disables the Temperature and Pressure Compensation (TPC) feature (Section 6.2.2). User Conc Units Change the concentration units of measure. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 71: Setup>Homescreen

    Setup>Homescreen menu or from Home page using the configuration icon (Figure 2-42). Figure 2-42. Homescreen Configuration An orientation to the Homescreen was presented in Section 2.4.1, including Figure 2-25 and Figure 2-26. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 72: Setup>Digital Outputs

    Go to the Utilities>Diagnostics>Digital Outputs menu to change the state (ON/OFF) of individual digital outputs. Figure 2-43. Digital Outputs Setup Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 73: Setup>Analog Outputs

    • noise around the zero point. Allow Overrange: check to allow a ± 5% over-range; uncheck to disable over-range • if the recording device is sensitive to excess voltage or current. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 74: Figure 2-45. Analog Outputs Group Calibration Screen

    While these are the physical limits of the current loop modules, typical applications use 2-20 mA or 4-20 mA for the lower and upper limits. For manual calibration adjustments, see Section 2.5.8.1 for voltage and Section 2.5.8.2 for current. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 75: Manual Calibration Of Voltage Range Analog Outputs

    0.1 VDC ±0.0005V 90 mV ±0.001V 0.02 mV 1 VDC ±0.001V 900 mV ±0.001V 0.24 mV 5 VDC ±0.002V 4500 mV ±0.003V 1.22 mV 10 VDC ±0.004V 4500 mV ±0.006V 2.44 mV 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 76: Manual Adjustment Of Current Range Analog Outputs

    Figure 2-4 for pin assignments and diagram of the analog output connector). This allows the use of a voltmeter connected across the resistor to measure converter output as VDC or mVDC. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 77: Setup>Instrument

    When an instrument is connected to a network that is connected to the Internet, follow the instructions on this Remote Update page to check for and activate software/firmware updates. (Also refer to Section 5.3). 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 78: Setup>Comm (Communications)

    Enable/disable the requirement for a password for this serial port to respond. The Security only command that is active is the request-for-help command (? CR). Stop bits Select either 0 or 1 stop bit (typically set in conjunction with Parity and Data bits). Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 79: Tcp Port1

    2.5.10.3 TCP PORT2 TCP Port2 is configured with the port number for MODBUS. 2.5.10.4 TCP PORT3 TCP Port3 is configured with the port number for Hessen. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 80: Network Settings

    A string of numbers very similar to the Instrument IP address (e.g. Default 192.168.76.1.) that is the address of the computer used by your LAN and serves Gateway as a router to access the Internet or another network. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 81: Transferring Configuration To Other Instruments

    7. When the Status field indicates that the USB drive has been detected, press the “Upload Configuration to Instrument” Start button. 8. When the Status field indicates that the upload is complete, remove the flash drive. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 82: Communications And Remote Operation

    RS232 on instrument rear panel) and/or COM2 (labeled COM2 on instrument rear panel) for communication modes, baud rate and serial communications. If using a USB option communication connection, setup requires that the instrument’s baud rate and personal computer baud rate match. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 83: Serial Communication: Rs-232

    A code-activated switch (CAS), can also be used on either port to connect typically between 2 and 16 send/receive instruments (host computer(s) printers, data loggers, analyzers, monitors, calibrators, etc.) into one communications hub. Contact Teledyne API Sales (front cover, this manual) for more information on CAS systems.
  • Page 84: Communications Protocols

    3.4.1 MODBUS These instructions assume that the user is familiar with MODBUS communications, and provide minimal information to get started. Please refer to the Teledyne API MODBUS manual, PN 06276, and to www.modbus.org for MODBUS communication protocols. Minimum Requirements: Instrument firmware with MODBUS capabilities installed •...
  • Page 85: Figure 3-2. Modbus Via Serial Communication (Example)

    Setup>Vars>Instrument ID menu. Next, for the settings to take effect, power off the analyzer, wait 5 seconds, and power it on again. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 86: Hessen

    Important instrument in the multidrop chain via the Setup>Vars>Instrument ID menu. The Hessen protocol is not strictly defined; therefore, while Teledyne API’s application is completely compatible with the protocol itself, it may be different from implementations by other companies. 3.4.2.1 HESSEN COM PORT CONFIGURATION...
  • Page 87: Hessen Settings Configuration

    TYPE 2 has more flexibility when operating with instruments that can measure more than one type of gas. For more specific information about the difference between the two versions, download the Manual Addendum for Hessen Protocol from the Teledyne API's web site: http://www.teledyne-api.com/manuals/.
  • Page 88: Table 3-3. Default Hessen Status Flag Assignments

    CANNOT DYN SPAN RELAY BOARD WARNING INVALID CONC REAR BOARD NOT DETECTED Instrument is in MP CAL mode Instrument is in MP CAL mode Only appears if the IZS option is installed Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 89: Calibration

    Have a dew point of -20°C Devices that condition ambient air by drying and removing any pollutants, such as the Teledyne API’ Model 701 Zero Air Module, are ideal for producing Zero Air. 083730400A DCN8130 Teledyne API – Model T400 O...
  • Page 90: Calibration (Span) Gas

    T400 successfully rejects interference from all of these with the exception of mercury vapor. If the Model T400 is installed in an environment where the presence of mercury vapor is suspected, steps should be taken to remove the mercury vapor from the sample gas before it enters the analyzer.
  • Page 91: Data Recording Devices

    Sample port, and check the reading after the Stability falls below 1.0 PPB (either in the gas graph or in the Dashboard). Otherwise, follow the steps presented in Sections 4.2.1.1 and 4.2.1.2. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 92: Zero Calibration Check And Actual Calibration

    In the Cal Result window, press OK. 3. Press the Stop button and return to Home screen. 4. In the Dashboard, check and record the Slope(s) and the Offset(s). (See Section 4.4 regarding these values). Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 93: Calibration And Check Procedures With Valve Options Installed

    Figure 4-2. Zero and Span Calibration Screens Follow the instructions in Section 4.2.1, except instead of the M-P menu, go to the Calibration>Zero menu for Zero cal and to the Calibration>Span menu for O3 Span cal. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 94: Use Of Zero/Span Valve With Remote Contact Closure

    (Zero, Low or High), or be disabled. Table 4-1 and Table 4-2 show how to set up the operating states of each calibration or check, and Table 4-3 shows how to program the execution of each. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 95: Figure 4-3. Auto Cal Page

    Calibrate and Enabled boxes are also High causes a High Span check when the Enabled box is also the Calibrate box is unchecked ( disables the High Span calibration and check. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 96: Table 4-2. Auto Cal Setup Combinations

    COM ports or the contact closure inputs will override any sequence that is in progress. It is recommended that calibrations be performed using • external sources of Zero Air and Span Gas whose accuracy is traceable to EPA standards. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 97: Calibration Quality

    Section 5. CALIBRATION OF THE ELECTRONIC SUBSYSTEMS There are several electronic characteristics of the T400 analyzer’s photometer that may occasionally need checking or calibration These calibrations include Dark Calibration, Pressure Cal and Flow Cal (and O3 Gen Cal when the O Generator option is installed), all located under the Utilities>Diagnostics menu.
  • Page 98: O 3 Gen Calibration

    Comply with Code of Federal Regulations, Title 40 (downloadable from the U.S. Government Publishing Office at http://www.gpo.gov/fdsys/) and with Quality Assurance Guidance documents (available website, http://www3.epa.gov/ttn/amtic/qalist.html). Give special attention to specific regulations regarding the use and operation of photometric ozone analyzers. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 99: Maintenance And Service

    Zero or Span Calibration, refer to Sections 4.2.1.1 and 4.2.1.2, respectively. Alternatively, use the Auto Cal feature described in Section 4.3 with the CALIBRATE attribute set to OFF (not enabled). 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 100: Table 5-1. T400 Maintenance Schedule

    Perform Flow 5.6.6 Flow Check Every 6 Months Check Annually or after Perform Leak Check any Maintenance 5.6.5 Leak Check or Repair Pneumatic Examine and Yes if As needed lines clean cleaned Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 101 This page intentionally left blank. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 102: Predicting Diagnostics

    • Pneumatic leaks – instrument & sample system • Pneumatics becoming contaminated/dirty Increasing • Dirty particulate filter SLOPE Span Cal • Pneumatic leaks – instrument & sample system • Contaminated calibration gas Decreasing Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 103: Operational Health Checks

    Figure 5-1. Report Generation Page The report can also be set to generate periodically and sent to a Web services “cloud” where it is available for viewing by Teledyne API technical support personnel. Set this function with two Vars: Setup>Vars>Upload Report to Cloud: set to True.
  • Page 104: Manual Reload/Update Procedures

    6. When complete, as indicated in the Status field, press the Done button, which replaces the Cancel button, and remove the flash drive. 7. Power off and restart the instrument to complete the new firmware installation. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 105: Instrument Display Calibration (For Earlier Instruments)

    Test the accuracy of the calibration by touching parts of the screen and see that the mouse pointer follows your touches. 9. If you press the CANCEL button, the calibration won’t be altered. Otherwise, press the ACCEPT button. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 106: Time Zone Changes

    6. After the Time Zone is implemented first (Steps 1 through 5), then other changes to the date and/or time can be made, and recycling the power is not necessary. Figure 5-5. Time Zone Change Requirements Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 107: Maintenance Procedures

    To change the filter: 1. Turn OFF the analyzer to prevent drawing debris into the instrument. 2. Open the T400’s hinged front panel and unscrew the knurled retaining ring on the filter assembly. Figure 5-6. Replacing the Particulate Filter 3.
  • Page 108: Rebuilding The Sample Pump

    4. Remove the old scrubber by disconnecting the 9/16” fitting at the top of the O generator tower, then removing the scrubber. 5. Install the new scrubber by reversing these instructions. IZS Zero Air Scrubber Figure 5-7. Replacing the IZS Zero Air Scrubber Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 109: Izs Desiccant Option

    If the reading is < 10 in-Hg, the pump is in good condition and there are no large • leaks. 4. Check the sample gas flow. If the flow is <10 cm /min and stable, there are no large leaks in the instrument’s • pneumatics. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 110: Pressure Leak Check

    8. Once the leak has been located and repaired, the leak-down rate should be < 1 in- Hg (0.4 psi) in 5 minutes after the pressure is shut off. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 111: Performing A Sample Flow Check

    CAUTION – General Safety Hazard Do not cause the tube to bind against the metal housings: the tube may break and cause serious injury. 9. Clean the tube only with de-ionized water. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 112: Adjusting The Optical Bench Uv Lamp

    3. Locate the UV DETECTOR GAIN ADJUST POT on the photometer assembly, shown below in Figure 5-8 Figure 5-8. Optical Bench – Lamp Adjustment/ Installation 4. Navigate to Utilities>Diagnostics>Analog Inputs>Photo Detector. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 113: Replacing The Optical Bench Uv Lamp

    12. Using an insulated pot adjustment tool, turn the UV DETECTOR GAIN ADJUST POT (Figure 5-8) until the Photo Detector signal value is as close as possible to 4600 mV. 13. Replace the cover on the analyzer. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 114: Adjusting Or Replacing The Izs Ozone Generator Option Suv Lamp

    The lamp should be free enough to be rotated in the assembly by grasping the lamp cable. 7. Power on the analyzer and allow it to stabilize for at least 20 minutes. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 115: Figure 5-10. Location Of O

    11. Finish screwing in the UV lamp holder so that it is snug but not excessively tight. 12. Replace the analyzer’s cover. 13. Perform a leak check (Section 5.6.5). 14. Perform an ozone generator calibration (Section 4.5.3). 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 116: Service And Troubleshooting

    • Check for gas flow problems such as clogged or blocked internal/external gas lines, damaged seals, punctured gas lines, a damaged / malfunctioning pumps, etc. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 117: Fault Diagnosis With Alerts

    • Blocked Sample Inlet/Gas Line Sample flow rate is < 500 SAMPLE FLOW WARN • Dirty Particulate Filter cc/min or > 1000 cc/min. • Leak downstream of Critical Flow Orifice • Failed Flow Sensor 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 118: Fault Diagnosis With Dashboard Functions

    Values outside these acceptable ranges indicate a failure of one or more of the analyzer’s subsystems. Functions whose values are still within the acceptable range but have significantly changed from the measurement recorded on the factory data sheet may also indicate a failure. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 119: Table 5-4. Dashboard Functions - Indicated Failures

    Contamination of the zero air or span gas supply Instrument is miscalibrated SLOPE Blocked gas flow Faulty sample pressure sensor (P1) or circuitry Bad/incorrect span gas concentration. Values outside range indicate OFFSET Contamination of the zero air supply 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 120: The Diagnostic Signal I/O Function

    If after 30 – 60 seconds neither the DS5 is flashing nor is the front panel display working, then the CPU is bad and must be replaced. Motherboard CPU Status LED Figure 5-11. CPU Status Indicator Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 121: I 2 C Bus Watchdog Status Led

    Faulty connectors/wiring between Motherboard, Valve Driver board (Red) Continuously OFF Circuit) or Relay PCA Failed/Faulty +5 VDC Power Supply (PS1) 5.7.4.3 O OPTION STATUS LED Figure 5-12. Relay PCA Status LEDs Used for Troubleshooting 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 122: Table 5-6. Relay Board Watchdog And Status Led Failure Indications

    Lamp Heater • Faulty connectors/wiring Only applies on analyzers with metal wool scrubber option installed. Only applies to instruments with calibration valve options installed. Only applies on analyzers with IZS options installed. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 123: Flow Problems

    Check to make sure that: • All calibrators/generators are turned on and working correctly. • Gas bottles are not empty or low. • Valves, regulators and gas lines are not clogged or dirty. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 124: Low Flow

    Too many analyzers on the manifold. This can cause either a slope or offset error • because ambient gas with its pollutants will dilute the zero or span gas. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 125: Non-Repeatable Zero And Span

    5.7.6.2 NON-REPEATABLE ZERO AND SPAN As stated earlier, leaks both in the T400 and in the external system are a common source of unstable and non-repeatable readings. • Check for leaks in the pneumatic systems as described in Section 5.6.3.
  • Page 126: Box Temperature

    Gen heater has failed. • • Check the resistance between pins 5 and 6 on the six-pin connector adjacent to the UV Lamp on the O Generator. • It should be approximately 5 Ohms. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 127: Subsystem Checkout

    5.7.9.1 AC MAINS CONFIGURATION The T400 analyzer’s electronic systems will operate with any of the specified power regimes. As long as system is connected to 100-120 VAC or 220-240 VAC at either 50 or 60 Hz it will turn on and after about 30 seconds show a front panel display.
  • Page 128: Dc Power Supply

    C bus can be verified by observing the behavior of D1 on the relay PCA. Assuming that the DC power supplies are operating properly, the I C bus is operating properly if D1 on the relay PCA is flashing. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 129: Touchscreen Interface

    The table below lists the control device associated with a particular function: • Table 5-9. Relay Board Control Devices CONTROL FUNCTION IN SOCKET DEVICE UV Lamp Heater Optional IZS O Gen Heater Optional Metal Wool Scrubber All Valves 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 130: Pressure/Flow Sensor Assembly

    • With flow stopped (photometer inlet disconnected or pump turned OFF) the voltage should be approximately 1V. • If the voltage is incorrect, the flow sensor S3 is bad, the board is bad or there is a leak upstream of the sensor. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 131: Motherboard

    5. Alternatively, turn on and off the output noting the voltage on the voltmeter. • It should vary between 0 volts for ON and 5 volts for OFF. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 132: Control Inputs - Remote Zero, Span

    5.7.12.1 GENERAL RS-232 TROUBLESHOOTING Teledyne API analyzers use the RS-232 communications protocol to allow the instrument to be connected to a variety of computer-based equipment. RS-232 has been used for many years and as equipment has become more advanced, connections between various types of hardware have become increasingly difficult.
  • Page 133: Troubleshooting The Photometer

    1. Unplug the cable connector at P1 on the lamp power supply and confirm that +15VDC is present between Pins 1 and 2 on the cable connector. 2. If this voltage is incorrect, check the DC test points on the relay PCA as described in Section 5.7.9.2 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 134: Troubleshooting The Izs Option's O Generator

    5.7.15.1 REPAIRING SAMPLE FLOW CONTROL ASSEMBLY The critical flow orifice is housed in the flow control assembly (Teledyne API P/N 001760400) located on the top of the optical bench. A sintered filter protects the jewel orifice so it is unusual for the orifice to need replacing, but if it does, or the filter needs...
  • Page 135: Replacing The Standard Reference O3 Scrubber

    3. The IZS zero air scrubber is attached to the brass elbow inlet fitting on the top of the generator assembly. See Figure 5-15. 4. Disconnect 1/4” Tube Fitting nut on O generator inlet fitting. 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 136: Replacing The Metal Wool Scrubber Option

    IZS O Generator Zero Air Scrubber Location 5.7.15.4 REPLACING THE METAL WOOL SCRUBBER OPTION Please contact Teledyne API’s Technical Support for this procedure. 5.7.15.5 REPLACING THE UV LAMP DRIVER BOARD Refer to Figure 5-16 for the following instructions: 1. Power off instrument and remove top.
  • Page 137: Figure 5-16: Uv Lamp Driver Board Replacement

    10. Reinsert driver board/communications power and lamp power supply connectors. 11. Return bench cover and screw into place. 12. Return cover to instrument. 13. Power on and check readings for accuracy. Figure 5-16: UV Lamp Driver Board Replacement 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 138: Disk-On-Module Replacement

    Press the chip all the way in. 8. Close the rear panel and turn on power to the machine. 9. If the replacement DOM carries a firmware revision, re-enter all of the setup information. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 139: Frequently Asked Questions

    FREQUENTLY ASKED QUESTIONS The following is a list from the Teledyne API’s Technical Support Department of the most commonly asked questions relating to this analyzer. QUESTION ANSWER This happens when the measured gas concentration differs significantly Why is the ZERO or SPAN button from the span or zero gas concentration value entered by the user.
  • Page 140: Technical Assistance

    Toll-free Phone: +1 800-324-5190 Phone: +1 858-657-9800 Fax: +1 858-657-9816 Email: api-techsupport@teledyne.com Website: http://www.teledyne-api.com/ Before contacting Teledyne API Tech Support, please fill out the problem report form, available online for electronic submission at http://www.teledyne-api.com. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 141: Theory Of Operation

    6. THEORY OF OPERATION The Model T400 ozone analyzer is a microprocessor-controlled analyzer that determines the concentration of Ozone (O ) in a sample gas drawn through the instrument. It requires that sample and calibration gasses be supplied at ambient atmospheric pressure in order to establish a stable gas flow through the absorption tube where the gas’...
  • Page 142: Calculating O Concentration

    (Euler’s number) is the absorption path, or the distance the light travels as it is being absorbed. is the concentration of the absorbing gas (in the case of the T400, Ozone (O α is the absorption coefficient that tells how well O absorbs light at the specific wavelength of interest.
  • Page 143: The Photometer Uv Absorption Path

    HOTOMETER BSORPTION IIn the most basic terms, the photometer of the Model T400 uses a high energy, mercury vapor lamp to generate a beam of UV light. This beam passes through a window of material specifically chosen to be both non-reactive to O and transparent to UV radiation at 254nm and into an absorption tube filled with Sample Gas.
  • Page 144: The Reference / Measurement Cycle

    O is present and when it is not. The Model T400 accomplishes this by alternately sending the sample gas directly to the absorption tube and passing it through a chemical scrubber that removes any O present.
  • Page 145: Interferent Rejection

    Measurement and Reference Phases rendering the analyzer useless for detecting O If the Model T400 is installed in an environment where the presence of mercury vapor is suspected, specific steps MUST be taken to remove the mercury vapor from the sample gas before it enters the analyzer.
  • Page 146: Sample Gas Air Flow

    6.2.1 S AMPLE The flow of sample gas through the T400 analyzer is produced by an internal pump that draws a small vacuum on the downstream side of a critical flow orifice thereby creating a controlled airflow through the analyzer’s absorption tube and other components. This requires the analyzer gas inlets be at or near ambient pressure usually managed by placing a vent line on the incoming gas line (see connections in Section 2.3.2).
  • Page 147: Particulate Filter

    ARTICULATE ILTER The T400 Ozone Analyzer comes equipped with a 47 mm diameter, Teflon, particulate filter with a 5 micron pore size. The filter is accessible through the front panel, which folds down to allow access, and should be changed according to the suggested maintenance schedule described in Table 5-1.
  • Page 148: Pneumatic Sensors

    The core of the analyzer is a microcomputer/central processing unit (CPU) that controls various internal processes, interprets data, makes calculations, and reports results using specialized firmware developed by Teledyne API. It communicates with the user as well as receives data from and issues commands to a variety of peripheral devices via a separate printed circuit assembly called the motherboard.
  • Page 149: Figure 6-5. Electronic Block Diagram

    Preamp Photometer Lamp Heater Photometer Absorption tube Optional IZS Photometer Lamp Power Generator Detector Supply Lamp Heater PHOTOMETER INTERNAL Optional Metal Wool Scrubber PUMP Thermocouple Sensor Figure 6-5. Electronic Block Diagram 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 150: Cpu

    The DOM is a 44-pin IDE flash drive with a storage capacity up to 128 MB. It is used to store the computer’s operating system, the Teledyne API firmware, and most of the operational data generated by the analyzer’s internal data acquisition system (DAS).
  • Page 151: Motherboard

    CPU. The A/D can be configured for several different input modes and ranges, but in the T400 it is used in uni-polar mode with a +5V full scale. The converter includes a 1% over and under-range.
  • Page 152: Analog Outputs

    This circuit monitors the +5V power supply during start-up and sets the analog outputs, external digital I/O ports, and I C circuitry to specific values until the CPU boots and the instrument software can establish control. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 153: I 6.3.3 Relay Board

    Figure 6-7. Relay Board Layout The most commonly used version of the Relay PCA installed in the T400 analyzer does not include the AC relays used in instruments where there are AC powered components requiring control. A plastic insulating safety shield covers the empty AC Relay sockets.
  • Page 154: Status Leds

    6.3.3.1 STATUS LEDS The LEDs located on the analyzer’s relay board show the current status on the various control functions performed by the relay board. They are listed on Table 6-1. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 155: I 2 C Watch Dog Circuitry

    6.3.3.3 HEATER CONTROL In the base version of the Model T400 photometric analyzer, there is only one DC heater operated by the relay PCA. It is attached to the Photometer UV Lamp housing and maintains the temperature of the UV Lamp at a constant 58ºC.
  • Page 156: Valve Control

    On instruments with the IZS valve option (OPT- 50G) one additional valve (the SAMPLE/CAL valve) selects either the sample inlet when the analyzer is in SAMPLE mode or the dry air inlet when the analyzer is in one of its calibration modes. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 157: Thermocouple Inputs And Configuration Jumper (Jp5)

    6.3.3.5 THERMOCOUPLE INPUTS AND CONFIGURATION JUMPER (JP5) In its base configuration, the T400 analyzer does not include any thermocouple sensors; however, in instruments where the optional metal wool scrubber (OPT-68) is installed, one thermocouple is used to sense the temperature of the scrubber. By default, this single thermocouple input is plugged into the TC1 input (J15) on the relay PCA.
  • Page 158: Power Supply/ Circuit Breaker

    PS 2 (+12 VDC) Optional IZS O POWER IN Generator UV Solenoid Drivers Lamp P/S COOLING OPTION MEASURE / FAN(S) VALVES REFERENCE (IZS, Sample/Cal, VALVES Zero/Span) Figure 6-12. Power Distribution Block Diagram Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 159: Ac Power Configuration

    OWER ONFIGURATION The T400 analyzer’s digital components will operate with any of the specified power regimes. As long as instrument is connected to 100-120 VAC or 220-240 VAC at either 50 or 60 Hz it will turn on and after about 30 seconds show a front panel display.
  • Page 160: Ac Configuration - Internal Pump (Jp7)

    Connects pump pin 1 to 220 / 240VAC power line 3 to 8 A jumper between pins 5 and 10 may be present on the jumper plug assembly, but is not functional on the T400. Figure 6-14. Pump AC Power Jumpers (JP7) 6.3.5.2...
  • Page 161: Photometer Layout And Operation

    Do not look at the UV lamp while the unit is operating. UV light can cause eye damage. Always use safety glasses made from UV blocking material when working with the UV Lamp Assembly. (Generic plastic glasses are not adequate). 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 162: Front Panel Touchscreen/Display Interface

    The LCD display is controlled directly by the CPU board. The touchscreen is interfaced to the CPU by means of a touchscreen controller that connects to the CPU via the internal USB bus and emulates a computer mouse. Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 163: Lvds Transmitter Board

    SOFTWARE OPERATION The instrument’s core module is a high performance, X86-based microcomputer running Windows CE. Inside Windows CE, special software developed by Teledyne API interprets user commands via the various interfaces, performs procedures and tasks, stores data in the CPU’s various memory devices and calculates the concentration of the sample gas.
  • Page 164: Adaptive Filter

    Unlike other analyzers that average the output signal over a fixed time period, the Model T400 averages over a set number of samples, where a new sample is calculated approximately every 3 seconds - this is a technique is known as boxcar averaging.
  • Page 165: Calibration - Slope And Offset

    The instrument slope and offset values recorded during the last calibration are viewable in the Dashboard (Setup>Dashboard to configure). 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 166 (Mbps) 100Base-T same as 10BaseT except ten times faster (100 Mbps) APICOM name of a remote control program offered by Teledyne-API to its customers ASSY Assembly Code-Activated Switch Corona Discharge, a frequently luminous discharge, at the surface of a conductor or between...
  • Page 167 Phase Lock Loop Photo Multiplier Tube, a vacuum tube of electrodes that multiply electrons collected and charged to create a detectable current signal P/N (or PN) Part Number Prevention of Significant Deterioration 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 168 Universal Serial Bus: a standard connection method to establish communication between peripheral devices and a host controller, such as a mouse and/or keyboard and a personal computer or laptop Variables, the variable settings of the instrument VARS Voltage-to-Frequency Zero / Span Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 169 Default Settings, 33 Control Inputs, 30, 47, 48, 92, 130, 150 Gas Filter Correlation, 142 Pin Assignments, 30 Gas Inlets, 115, 116, 154 Sample, 23 Control InputS Span, 23 Electrical Connections, 30 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 170 ENTR, 131 Pump EXIT, 131 Metal Wool Scrubber, 120, 152, 153, 158 Sample, 115, 158 AC Power, 158, 159 Troubleshooting, 123 Motherboard, 74, 115, 119, 125, 146, 157 MR Ratio, 100 Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 171 UV Absorption Path, 141, 142 Multidrop, 34, 36 Slope, 115 UV detector, 110, 115, 159 SLOPE, 100, 117 UV Lamp, 131, 159 SPAN CAL, 130 UV Light, 131, 159 Remote, 30 083730400A DCN8130 Teledyne API – Model T400 O Analyzer...
  • Page 172 Internal Pneumatics, 48 Zero/Span/Shutoff Valves Internal Pneumatics, 48 Pneumatic Set Up, 44 Zero/Span/Shutoff Valves VARS Menu Pneumatic Set Up, 44 Variable Names PHOTO_LAMP, 120, 124 Ventilation Clearance, 20 Venting Exhaust Line, 40, 43 Teledyne API – Model T400 O Analyzer 083730400A DCN8130...
  • Page 173 Teledyne API - Models T400 and 400E (04402G DCN6874) MODBUS Register Map MODBUS Description Units Register Address (dec., 0-based) MODBUS Floating Point Input Registers (32-bit IEEE 754 format; read in high-word, low-word order; read-only) Photometer detector measure reading Photometer detector reference reading...
  • Page 174 APPENDIX A-7: MODBUS Register Map Teledyne API - Models T400 and 400E (04402G DCN6874) MODBUS Description Units Register Address (dec., 0-based) External analog input 2 slope eng unit /V External analog input 2 offset eng unit External analog input 3 value...
  • Page 175 Teledyne API - Models T400 and 400E (04402G DCN6874) APPENDIX A-7: MODBUS Register Map MODBUS Description Units Register Address (dec., 0-based) Sample pressure warning System reset warning Rear board communication warning Relay board communication warning generator or photometer lamp I...
  • Page 176 This page intentionally left blank. 083730400A DCN8130...
  • Page 177 Appendix B - Wiring Diagram 083730400A DCN8130...

Table of Contents