T100 Instruction Manual
Customer: _______________________________________
Contact Name: ___________________________________
Site Address: _____________________________________________________________________________________
Serial No.: _______________________________________
1. Are there any failure messages? ____________________________________________________________________
________________________________________________________________________________________________
Please complete the following table: (Note: Depending on options installed, not all test parameters shown below will be
available in your instrument)
* If option is installed
PARAMETER
RANGE
STABIL
SAMP PRESS
SAMPLE FLOW
PMT SIGNAL WITH ZERO
AIR
PMT SIGNAL AT SPAN
GAS CONC
NORM PMT AT SPAN
GAS CONC
UV LAMP
LAMP RATIO
STR. LGT
DARK PMT
DARK LAMP
SLOPE
OFFSET
HVPS
RCELL TEMP
BOX TEMP
PMT TEMP
IZS TEMP*
ETEST
OTEST
REF_4096_MV
REF_GND
2. What is the SAMPLE FLOW and SAMPLE PRESSURE with the sample inlet on rear of machine capped?
SAMPLE FLOW: __________________ cc/min
PHONE: (303) 792-3300
068070000 Rev A
Warranty/Repair
Questionnaire T100
RECORDED VALUE
Values are in the Signal I/O
SAMPLE PRESSURE: ________________ In-Hg-A
TELEDYNE ML CUSTOMER SERVICE
EMAIL:
tml_support@teledyne.com
TOLL FREE: (800) 846-6062 FAX: (303) 799-4853
Phone: _______________________________________
Fax No: ______________________________________
Firmware Revision: _____________________________
UNITS
PPB/PPM
50 PPB to 20 PPM Standard
PPB
In-Hg-A
cc/min
mV
mV
PPB/PPM
mV
PPB/PPM
mV
%
PPB
mV
mV
mV
V
ºC
ºC
ºC
ºC
mV
mV
mV
4096mv±2mv and must be Stable
mV
0± 0.5 and must be Stable
ACCEPTABLE VALUE
1 PPB with Zero Air
Ambient ± 2 In-Hg-A
650 ± 10%
-20 to 150 mV
0-5000 mV
0-20000 PPB
0-5000 mV
0-20000 PPB
1000 to 4800 mV
30 to 120%
≤ 100 PPB/ Zero Air
-50 to 200 mV
-50 to 200 mV
1.0 ± 0.3
< 250 mV
≈ 400 – 900
50ºC ± 1
Ambient + ~ 5
7ºC ± 2º Constant
50ºC ± 1
2000 mV ± 1000
2000 mV ± 1000
Pg: C-1