Teledyne ML - T100 UV Fluorescence SO2 Analyzer
Troubleshooting & Repair
2. Remove the sample chamber from the PMT lens and filter housing by unscrewing the
4 hex screws that fasten the chamber to the housing.
3. Remove the four lens cover screws.
Figure 10-10:
UV Lens Housing / Filter Housing
4. Remove the lens/filter cover.
5. Carefully remove the PMT lens and set it aside on soft, lint-free cloth.
6. Remove the 3-piece, lens/filter spacer.
7. Carefully remove the PMT filter and set it aside on soft, lint-free cloth.
Figure 10-11: PMT UV Filter Housing Disassembled
06807A(DCNxxxx)