Operating Instructions
4.8.4.2. ANALOG OUTPUT CALIBRATION MODE
Exit at any time
to return to the
main DIAG
menu
If any of the channels have
not been calibrated this
message will read NO.
108
Analog output calibration should to be carried out on first startup of the analyzer
(performed in the factory as part of the configuration process) or whenever recalibration
is required. The analog outputs can be calibrated automatically, either as a group or
individually, or adjusted manually.
In its default mode, the instrument is configured for automatic calibration of all
channels, which is useful for clearing any analog calibration warnings associated with
channels that will not be used or connected to any input or recording device, e.g.,
datalogger.
Manual calibration should be used for the 0.1V range or in cases where the outputs must
be closely matched to the characteristics of the recording device. Manual calibration
requires the AUTOCAL feature to be disabled.
To calibrate the outputs as a group, activate the ANALOG I/O CONFIGURATION
MENU from the DIAG Menu (refer to Figure 4-17), then press:
DIAG
ANALOG I / O CONFIGURATION
PREV
NEXT
DIAG AIO
AOUTS CALIBRATED:
< SET SET>
CAL
DIAG AIO
AUTO CALIBRATING CONC_OUT_1
(continues for any active channels)
AUTO CALIBRATING TEST_OUTPUT
DIAG AIO
AOUTS CALIBRATED:
< SET SET>
CAL
DIAG – Analog Output Calibration Mode
Figure 4-21:
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
ENTR
EXIT
NO
EXIT
YES
EXIT
If AutoCal has been
turned off for any
channel, the message
for that channel will be
similar to:
NOT AUTO CAL
CONC_OUT_1
Exit to return to
the I/O
configuration
menu