4.8.4. ANALOG I/O CONFIGURATION

SUB MENU
AOUTS CALIBRATED:
CONC_OUT_1
CONC_OUT_2
TEST OUTPUT
CONC_OUT_3
AIN CALIBRATED
XIN1
.
.
.
XIN8
Table 6-8 lists the analog I/O functions that are available in the T100.
Table 4-9: DIAG - Analog I/O Functions
Shows the status of the analog output calibration (YES/NO) and initiates a calibration of all
analog output channels.
Sets the basic electronic configuration of the A1 analog output (SO
RANGE: Selects the signal type (voltage or current loop) and full scale level of the
output.
REC_OFS: Allows setting a voltage offset (not available when RANGE is set to Current
Loop (CURR).
AUTO_CAL: Performs the same calibration as AOUT CALIBRATED, but on this one
channel only.
NOTE: Any change to RANGE or REC_OFS requires recalibration of this output.
Same as for CONC_OUT_1 but for analog channel 2 (SO
Same as for CONC_OUT_1 but for analog channel 3 (TEST)
(Not available in the analyzer's standard configuration; applies when optional sensor installed).
Shows the calibration status (YES/NO) and initiates a calibration of the analog input channels.
For each of 8 external analog inputs channels, shows the gain, offset, engineering units, and
whether the channel is to show up as a Test function.
Table 4-10: Analog Output Voltage Ranges
RANGE
MINIMUM OUTPUT
0-0.1 V
0-1 V
0-5 V
0-10 V
The default offset for all ranges is 0 VDC.
FUNCTION
MAXIMUM OUTPUT
-5 mV
-0.05 V
-0.25 V
-0.5 V
Operating Instructions
). There are three options:
2
)
2
+105 mV
+1.05 V
+5.25 V
+10.5 V
105