Teledyne ML - T100 UV Fluorescence SO2 Analyzer
The SO
span concentration
2
values automatically default
to 450.0 Conc.
To change this value to the
actual concentration of the
span gas, enter the number
by pressing the key under
each digit until the expected
value appears.
The span gas concentration
should always be 90% of the
selected reporting range
EXAMPLE
Reporting range = 800 ppb
Span gas conc.= 720 ppb
Figure 6-6: Setup for Manual Calibration with Z/S Valve Option Installed (Step 2)
STEP TWO: Set the expected SO
SAMPLE
RANGE = 500.000 PPB
< TST TST >
CAL
M-P CAL
RANGE = 500.000 PPB
< TST TST >
ZERO
M-P CAL
SO2 SPAN CONC: 450.0 Conc
0
0
0
span gas
value:
2
SO2 =XXX.X
SETUP
SO2 =XXX.X
CONC
EXIT
4
5
.0
ENTR EXIT
Calibration Procedures
This sequence causes the
analyzer to prompt for the
expected SO
span
2
concentration.
EXIT ignores the new setting
and returns to the previous
display.
ENTR accepts the new setting
and returns to the
previous display..