Table 3-11: Izs Valve Operating States - Teledyne T100 Instruction Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for T100:
Table of Contents

Advertisement

Teledyne ML - T100 UV Fluorescence SO2 Analyzer
Manually from the analyzer's front panel by using the SIGNAL I/O controls under the
DIAG Menu (refer to Section 4.8.2),
By activating the instrument's AutoCal feature (refer to Section 6.8),
Remotely by using the external digital control inputs (refer to Section 5.3.1.2 and Section
6.7.1),
Remotely through the RS-232/485 serial I/O ports (refer to Appendix A-6 for the applicable
commands), or
Remotely via Ethernet
IMPORTANT
3.5.4.1. PERMEATION TUBE HEATER
3.5.4.2. SPAN GAS CONCENTRATION VARIATION
Size of the membrane: The larger the area of the membrane, the more permeation occurs.
Temperature of the SO
tube and therefore increases the effusion rate.
Table 3-11 describes the operational state of the valves associated with the IZS option
during the analyzer's various operating modes.

Table 3-11: IZS Valve Operating States

MODE
SAMPLE
ZERO CAL
SPAN CAL
The state of the IZS valves can also be controlled by any of the following means:
The permeation tube is not included in the IZS Option and must be ordered separately.
Refer to Section 1.3 for permeation tube options.
In order to keep the permeation rate constant, the IZS enclosure is heated to a constant
50 C (10° above the maximum operating temperature of the instrument). The IZS heater
is controlled by a precise PID (Proportional/Integral/Derivative) temperature control
loop. A thermistor measures the actual temperature and reports it to the CPU for control
feedback.
The IZS option includes an external zero air scrubber assembly that removes all SO
zero air source. The scrubber is filled with activated charcoal.
Span gas is created when zero air passes over a permeation tube containing liquid SO
under high pressure, which slowly permeates through a PTFE membrane into the
surrounding air. The speed at which the SO
effusion rate. The concentration of the span gas is determined by three factors:
: Increasing the temperature of the increases the pressure inside the
2
VALVE
Sample/Cal
Open to SAMPLE inlet
Zero/Span
Open to ZERO AIR inlet
Sample/Cal
Open to zero/span valve
Zero/Span
Open to ZERO AIR inlet
Sample/Cal
Open to zero/span valve
Zero/Span
Open to SPAN GAS inlet
permeates the membrane is called the
2
Getting Started
CONDITION
the
2
2
63