Teledyne T100 Instruction Manual page 148

Uv fluorescence so2 analyzer
Hide thumbs Also See for T100:
Table of Contents

Advertisement

Remote Operation & Advanced Features
Use PREV and NEXT to
scroll to the data channel
to be edited.
Set the number of days
between reports (0-365).
Press buttons to set amount of
time between reports, in hours
(HH) and/or minutes (MM)
(max: 23:59). 01:00 sets a
report to be made every hour.
5.2.2.6. NUMBER OF RECORDS
148
To define the REPORT PERIOD, follow the instruction shown in Figure 5-11, then
press:
Edit Data Channel Menu
SETUP X.X
0) CONC:
PREV NEXT
SETUP X.X
NAME:CONC
<SET
SET>
EDIT
Press SET> until you reach REPORT PERIOD ...
SETUP X.X
REPORT PERIOD:000:00:05
<SET
SET>
EDIT
SETUP X.X
REPORT PERIODD:DAYS:0
0
0
0
SETUP X.X
REPORT PERIODD:TIME:01:00
0
1
0
IIf at any time an invalidl entry is selected (e.g., days > 366)
the ENTR button will disappear from the display.
DAS – Define the Report Period
Figure 5-16:
The number of data records in the T100 is limited to about a cumulative one million data
points in all channels (one megabyte of space on the disk-on-module). However, the
actual number of records is also limited by the total number of parameters and channels
and other settings in the DAS configuration. Every additional data channel, parameter,
number of samples setting etc. will reduce the maximum amount of data points
somewhat. In general, however, the maximum data capacity is divided amongst all
channels (max: 20) and parameters (max: 50 per channel).
The DAS will check the amount of available data space and prevent the user from
specifying too many records at any given point. If, for example, the DAS memory space
can accommodate 375 more data records, the ENTR button will disappear when trying
to specify more than that number of records. This check for memory space may also
make an upload of a DAS configuration with APICOM or a Terminal program fail, if
the combined number of records would be exceeded. In this case, it is suggested to
either try from the front panel what the maximum number of records can be or use trial-
and-error in designing the DAS script or calculate the number of records using the DAS
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
ATIMER,
2,
4032, R
INS
DEL
EDIT
PRNT
EXIT
PRINT
EXIT
PRINT
EXIT
ENTR
EXIT
0
ENTR
EXIT
Exits to the main
Data Acquisition
menu.
ENTR accepts the new string and
returns to the previous menu.
EXIT ignores the new string and
returns to the previous menu.