APPENDIX A-3: Warnings and Test Functions

Table A-2:
NAME
WSYSRES
SYSTEM RESET
WDATAINIT
DATA INITIALIZED
WCONFIGINIT
CONFIG INITIALIZED
7
WSO2ALARM1
SO2 ALARM 1 WARN
7
WSO2ALARM2
SO2 ALARM 2 WARN
10+7
WO2ALARM1
O2 ALARM 1 WARN
10+7
WO2ALARM2
O2 ALARM 2 WARN
11+7
WCO2ALARM1
CO2 ALARM 1 WARN
11+7
WCO2ALARM2
CO2 ALARM 2 WARN
WPMT
PMT DET WARNING
WUVLAMP
UV LAMP WARNING
WSAMPFLOW
SAMPLE FLOW WARN
WSAMPPRESS
SAMPLE PRESS WARN
5
WVACPRESS
VACUUM PRESS WARN
WBOXTEMP
BOX TEMP WARNING
WRCELLTEMP
RCELL TEMP WARNING
10
WO2CELLTEMP
O2 CELL TEMP WARN
WIZSTEMP
IZS TEMP WARNING
WPMTTEMP
PMT TEMP WARNING
WDARKCAL
DARK CAL WARNING
WHVPS
HVPS WARNING
WDYNZERO
CANNOT DYN ZERO
WDYNSPAN
CANNOT DYN SPAN
WREARBOARD
REAR BOARD NOT DET
WRELAYBOARD
RELAY BOARD WARN
Warning Messages, Software Version G.4
MESSAGE TEXT
05036 Rev D
APPENDIX A - Version Specific Software Documentation
DESCRIPTION
Warnings
Instrument was power-cycled or the CPU was reset.
Data storage was erased.
Configuration storage was reset to factory configuration or
erased.
SO
concentration alarm limit #1 exceeded
2
SO
concentration alarm limit #2 exceeded
2
O
concentration alarm limit #1 exceeded
2
O
concentration alarm limit #2 exceeded
2
CO
concentration alarm limit #1 exceeded
2
CO
concentration alarm limit #2 exceeded
2
PMT detector outside of warning limits specified by
DETECTOR_LIMIT variable.
UV lamp reading outside of warning limits specified by
DETECTOR_LIMIT variable.
Sample flow outside of warning limits specified by
SAMP_FLOW_SET variable.
Sample pressure outside of warning limits specified by
SAMP_PRESS_SET variable.
Vacuum pressure outside of warning limits specified by
VAC_PRESS_SET variable.
Chassis temperature outside of warning limits specified by
BOX_SET variable.
Reaction cell temperature outside of warning limits specified by
RCELL_SET variable.
O
sensor cell temperature outside of warning limits specified by
2
O2_CELL_SET variable.
IZS temperature outside of warning limits specified by IZS_SET
variable.
PMT temperature outside of warning limits specified by
PMT_SET variable.
Dark offset above limit specified by DARK_LIMIT variable.
High voltage power supply output outside of warning limits
specified by HVPS_SET variable.
Contact closure zero calibration failed while DYN_ZERO was set
to ON.
Contact closure span calibration failed while DYN_SPAN was set
to ON.
Rear board was not detected during power up.
Firmware is unable to communicate with the relay board.
A-21