Figure 3-27: So 2 Span Gas Setting - Teledyne T100 Instruction Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for T100:
Table of Contents

Advertisement

Getting Started
3.7.1.3. SET SO
2
Only appears if either
the optional O
or CO
2
sensors are installed.
The SO
span concentration value
2
is automatically default to
400.0 PPB.
If this is not the the concentration of
the span gas being used, toggle
these buttons to set the correct
concentration of the SO
calibration
2
gas.
74
SPAN GAS CONCENTRATION
Set the expected SO
span gas concentration. This should be 80% of the concentration
2
range for which the analyzer's analog output range is set.
SAMPLE
RANGE=500.0 PPB
<TST
TST>
CAL
SAMPLE
CO
O2
2
SAMPLE
LOW
HIGH
M-P CAL
RANGE=400.0 PPB
<TST TST> ZERO SPAN
M-P CAL
SO2 SPAN CONC:400.0 Conc
0
4
Figure 3-27:
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
SO2= XXXX
GAS TO CAL:CO
ENTR
RANGE TO CAL:LOW
ENTR
SO2= XXXX
CONC
0
0
.0
0
ENTR EXIT
SO
Span Gas Setting
2
SETUP
EXIT
Only appears if either
the analyzer is set for
EXIT
DUAL range mode.
EXIT
EXIT ignores the new
setting and returns to
the previous display.
ENTR accepts the new
setting and returns to
the
CONCENTRATION
MENU.