Calibration Procedures
The SPAN key now appears
during the transition from
zero to span.
You may see both keys.
Figure 6-3: Setup for Manual Calibration without Z/S valve or IZS Option (Step 3)
176
SAMPLE
RANGE = 500.0 PPB
< TST
TST
>
CAL
SAMPLE
RANGE = 500.0 PPB
< TST TST >
CAL
ACTION:
Allow zero gas to enter the sample port at the
rear of the instrument.
M-P CAL
SO2 STB=X.XXX PPB
< TST TST >
CAL
M-P CAL
SO2 STB=X.XXX PPB
< TST TST >
ZERO
CONC
M-P CAL
SO2 STB=X.XXX PPB
< TST TST >
ENTR
CONC
ACTION:
Allow span gas to enter the sample port at the
rear of the instrument.
M-P CAL
SO2 STB=X.XXX PPB
< TST TST >
SPAN
CONC
M-P CAL
RANGE = 500.0 PPB
< TST TST >
ENTR
SPAN CONC
M-P CAL
RANGE = 500.0 PPB
< TST TST > ENTR
CONC
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
SO2 =XXX.X
SETUP
SO2 =XXX.X
SETUP
SO2
=XXX.X
SETUP
SO2
=XXX.X
EXIT
SO2
=XXX.X
EXIT
SO2
=XXX.X
EXIT
SO2
=XXX.X
EXIT
SO2
=XXX.X
EXIT
Set the Display to show the
SO2 STB test function.
This function calculates the
stability of the SO
2
measurement
Wait until SO2 STB
falls below 0.5 ppb.
This may take several
minutes.
Press ENTR to changes the
OFFSET & SLOPE values for the
SO
measurements.
2
Press EXIT to leave the calibration
unchanged and return to the
previous menu.
The value of
SO2 STB may jump
significantly.
Wait until it falls back
below 0.5 ppb.
This may take several
minutes.
Press ENTR to change the
OFFSET & SLOPE values for the
SO
measurements.
2
Press EXIT to leave the calibration
unchanged and return to the
previous menu.
EXIT returns to the main
SAMPLE display