Level 1 Calibrations Versus Level 2 Checks - Teledyne T100 Instruction Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for T100:
Table of Contents

Advertisement

Teledyne ML
- T100 UV Fluorescence SO2 Analyzer
Analyzer operational
settings
Analyzer operational
check
Precision check

7.2. LEVEL 1 CALIBRATIONS VERSUS LEVEL 2 CHECKS

TEST measurements at
nominal values
2. T100 in SAMPLE mode
Zero and span within
tolerance limits as described
in Subsec. 9.1.3 of Sec.
2.0.9 (Q.A. Handbook)
Assess precision as
described in Sec. 2.0.8 and
Subsec. 3.4.3 (Ibid.)
Essential to quality assurance are scheduled checks for verifying the operational status
of the monitoring system. The operator should visit the site at least once each week. It is
recommended Level 1 zero and span check conducted on the analyzer every two weeks.
Level 2 zero and span checks should be conducted at a frequency desired by the user.
Definitions of these terms are given in Table 7-4.
In addition, an independent precision check between 0.08 and 0.10 ppm must be carried
out at least once every two weeks. Table 7-3 summarizes the quality assurance activities
for routine operations. A discussion of each activity appears in the following sections.
To provide for documentation and accountability of activities, a checklist should be
compiled and then filled out by the field operator as each activity is completed.
Weekly visual inspection
Level 1 zero/span every 2
weeks; Level 2 between Level
1 checks at frequency desired
analyzer by user
Every 2 weeks, Subsec. 3.4.3
(Ibid.)
EPA Protocol Calibration
Adjust or repair as needed
Find source of error and repair
After corrective action, re-
calibrate analyzer
Calc, report precision, Sec. 2.0.8
(Ibid.)