Calibration Procedures
Analyzer enters ZERO
CAL mode.
Analyzer enters SPAN
CAL Mode.
If either the ZERO or
SPAN button fails to
appear, see Chapter 11
for troubleshooting tips.
Figure 6-7: Setup for Manual Calibration with Z/S Valve Option Installed (Step 3)
180
Step Three: Perform the calibration or calibration check according to the following
flow chart:
SAMPLE
RANGE = 500.000 PPB
< TST
TST >
CAL
SAMPLE
< TST TST > CAL
Allow zero gas to enter the sample port at the
ZERO CAL M
< TST TST >
ZERO
ZERO CAL M
< TST TST >
ENTR
Pressing ENTR changes the calibration of the instrument.
ZERO CAL M
< TST TST > ZERO
ZERO CAL M
< TST TST > CAL CALZ
SPAN CAL M
< TST TST >
SPAN
SPAN CAL M
< TST TST >
ENTR
Pressing ENTR changes the calibration of the instrument.
SPAN CAL M
< TST TST >
SPAN
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
SO2 =XXX.X
CALZ CALS
SETUP
STABIL=XXX.X PPB
SO2 =XXX.X
CALZ
CALS
SETUP
ACTION:
rear of the instrument.
STABIL=XXX.X PPB
SO2 =XXX.X
CONC
STABIL=XXX.X PPB
SO2 =XXX.X
CONC
EXIT
EXIT
STABIL=XXX.X PPB
SO2 =XXX.X
CONC
EXIT
STABIL=XXX.X PPB
SO2=X.XX
CALS
SETUP
STABIL=XXX.X PPB
SO2 =XXX.X
CONC
EXIT
STABIL=XXX.X PPB
SO2 =XXX.X
CONC
EXIT
STABIL=XXX.X PPB
SO2 =XXX.X
CONC
EXIT
Scroll the display to the
STABIL test function. This
function calculates the stability
of the SO
measurements.
2
Wait until STABIL
falls below 0.05
ppb. This may
take several
minutes.
EXIT returns the unit to
SAMPLE mode without
changing the calibration
values.
X
The value of STABIL
may jump
significantly. Wait
until it falls below 0.5
ppb. This may take
several minutes.
EXIT returns to the
SAMPLE mode without
changing the calibration
values.
EXIT returns to the
main SAMPLE
display