Figure 2-46. Analog Outputs Manual Calibration Screen (AOUT2 Example)

Table 2-11. Analog Output Voltage/Current Range
RANGE
100mV
1V
5V
10V
Current
1
Each range is usable from -5% to +5% of the rated span.
2
While these are the physical limits of the current loop modules, typical applications use 2-20 mA or 4-20
mA for the lower and upper limits.
For manual calibration adjustments, see Section 2.5.8.1 for voltage and Section 2.5.8.2
for current.
72

Figure 2-45. Analog Outputs Group Calibration Screen

RANGE SPAN
1
0-100 mVDC
0-1 VDC
0-5 VDC
0-10 VDC
2
0-20 mA
Teledyne API – Model T400 O
MINIMUM OUTPUT
MAXIMUM OUTPUT
-5 mVDC
-0.05 VDC
-0.25 VDC
-0.5 VDC
0 mA
Analyzer
3
105 mVDC
1.05 VDC
5.25 VDC
10.5 VDC
20 mA
083730400A DCN8130