Relay Board Status Leds; Gas Flow Problems; Zero Or Low Sample Flow - Teledyne T100 Instruction Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for T100:
Table of Contents

Advertisement

Troubleshooting & Repair

10.2.3. RELAY BOARD STATUS LEDS

LED
D1

10.3. GAS FLOW PROBLEMS

Flow is too high
Flow is greater than zero, but is too low, and/or unstable
Flow is zero (no flow)

10.3.1. ZERO OR LOW SAMPLE FLOW

Check for actual sample flow
Check pressures
Carry out a leak check
268
The most important status LED on the relay board is the red I
labeled D1 (or W/D), which indicates the health of the I
LED is located in the upper left-hand corner of the relay board when looking at the
electronic components. If D1 is blinking, then the other LEDs can be used in conjunction
with the DIAG menu I/O functions to test hardware functionality by switching devices on
and off and watching the corresponding LED turn on or off. The LED only indicates that
the logic signal for an output has been activated. If the output driver (i.e. the relay or
valve driver IC) is defective, then the LED will light up, but the attached peripheral
device will not turn on.
Table 10-3: Relay Board Status LEDs
COLOR
FUNCTION
Watchdog Circuit;
red
2
I
C bus operation.
The standard analyzer has one main flow path. With the IZS option installed, there is a
second flow path through the IZS oven that runs whenever the IZS is on standby to purge
SO
from the oven chamber. The IZS flow is not measured so there is no reading for it on
2
the front panel display. The full flow diagrams of the standard configuration (refer to
Figure 3-18) and with options installed (refer to Figure 3-19 and Figure 3-20) help in
troubleshooting flow problems. In general, flow problems can be divided into three
categories:
When troubleshooting flow problems, it is essential to confirm the actual flow rate
without relying on the analyzer's flow display. The use of an independent, external flow
meter to perform a flow check as described in Section 10.5.2 is essential.
If the pump is operating but the unit reports a XXXX gas flow, do the following three
steps:
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
FAULT
STATUS
Failed/Halted CPU
Faulty Mother Board, Valve Driver board or
Relay PCA
Continuously
ON or OFF
Faulty Connectors/Wiring between Motherboard,
Valve Driver board or Relay PCA
Failed/Faulty +5 VDC Power Supply (PS1)
06807A(DCNxxxx)
2
C Bus watch-dog LED,
2
C communications bus. This
INDICATED FAILURE(S)