4.8.6. ELECTRICAL TEST

The electrical test function creates a current, which substitutes the PMT signal, and
feeds it into the preamplifier board. This signal is generated by circuitry on the pre-
amplifier board itself and tests the filtering and amplification functions of that assembly
along with the A/D converter on the motherboard. It does not test the PMT itself. The
electrical test should produce a PMT signal of about 2000 ±1000 mV.
Access the Electrical Test from the DIAG Menu (refer to Figure 4-17), then press:
While the electrical test is
activated, PMT should equal:
2000 mV ± 1000 mV
DIAG
PREV
NEXT
JUMP
DIAG
PREV NEXT
DIAG ELEC
RANGE = 500.0 PPB
<TST TST>
DIAG ELEC
PMT = 1732 MV
<TST TST>
DIAG – Electrical Test
Figure 4-32:
Operating Instructions
SIGNAL I / O
ENTR
EXIT
Press NEXT until...
ELECTRICAL TEST
ENTR
SO2 =XXX.X
Press TST until...
SO2=X.XXX
EXIT
EXIT
EXIT